Dear Friends,
I need your support on a Toshiba XPRESS / GX CT.
When I visited the site I got an error:
Overlong exposure time error in the XC
Contact service personnel
With details:
04 : E 0 : Error occurred in the high-voltage generator
09 : E64 : Low mA
24 : E 0 : XC error
27 : E 3 :Exposure timeout error
On XP board:
LED 5- âONâ: Name: LOWMA, Description: no current output from the anode is detected
LED 9- âONâ: Name: H_ERR, Description: A protection circuit is operating or control of XC is abnormal
I found out defective elements and replaced them
-F 28 / 100 A broken
-One IGBT in inverter part with broken collector / gate section. I replaced both IGBT and one Diode.
XC log did not helped.
The error is still present.
I have seen this error with a broken X-ray focus.
Further troubleshooting- Replacement of:
- XC board
- XP board
- X-ray tube
- HV transformer
- Check HV cables and other elements.
Error is still present no matter what action is taken.
The CT gives you the possibility to make an exposure- you can choose warming up or DCA scan- X-ray tube anode is accelerating, X-ray warning lamp is ON, start button is ON, but when you push the start button you hear the main contactor cut off the 290 V DC and no exposure is performed.
Obviously the problem is in HV generator.
I need support on HV generator protection circuits and interlocks.
Any new idea will be highly appreciated.
Thanks in advance
Hi, Normally when IGBT Fails, it can damage the PCB of Gate Trigger circuits. Hope you have verified this already. Please send the picture of IGBT and Trigger circuits to shiju@tecsila.com or Whatsapp +91 94440 26607.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Kostadin Petrov
CT XPRESS/GX
July 18, 2017 05:10
Dear Friends,
I need your support on a Toshiba XPRESS / GX CT.
When I visited the site I got an error:
Overlong exposure time error in the XC
Contact service personnel
With details:
04 : E 0 : Error occurred in the high-voltage generator
09 : E64 : Low mA
24 : E 0 : XC error
27 : E 3 :Exposure timeout error
On XP board:
LED 5- âONâ: Name: LOWMA, Description: no current output from the anode is detected
LED 9- âONâ: Name: H_ERR, Description: A protection circuit is operating or control of XC is abnormal
I found out defective elements and replaced them
-F 28 / 100 A broken
-One IGBT in inverter part with broken collector / gate section. I replaced both IGBT and one Diode.
XC log did not helped.
The error is still present.
I have seen this error with a broken X-ray focus.
Further troubleshooting- Replacement of:
- XC board
- XP board
- X-ray tube
- HV transformer
- Check HV cables and other elements.
Error is still present no matter what action is taken.
The CT gives you the possibility to make an exposure- you can choose warming up or DCA scan- X-ray tube anode is accelerating, X-ray warning lamp is ON, start button is ON, but when you push the start button you hear the main contactor cut off the 290 V DC and no exposure is performed.
Obviously the problem is in HV generator.
I need support on HV generator protection circuits and interlocks.
Any new idea will be highly appreciated.
Thanks in advance
to rate and post a comment
Shiju Karakkoottil
re: CT XPRESS/GX
July 21, 2017 10:25
Hi, Normally when IGBT Fails, it can damage the PCB of Gate Trigger circuits. Hope you have verified this already. Please send the picture of IGBT and Trigger circuits to shiju@tecsila.com or Whatsapp +91 94440 26607.
to rate and post a comment