CTe artefact image

April 11, 2013 12:02

hi TEAM CTe, Hispeed
I have a problem quality image not stable, if I make a simple warmup without calibration, I have a cible artefact, I makes then calibration dayli and the artefact disappears, but one hour later, artefact reappears, after recalibration dayly it no artefact.
in addition, I have the aircal drift transistor in red in the small service home.
perhaps this has a report.
Another thing, before reloading a LFW due to nonrecon image, also replaced tube X-ray broken, the aircal drift were green, and I also have since error HINV and NPR error.
I type in a window Shell, hwhinv to remove error inventoty but that resoud nothing. Have ideas to help me. Thank you in advance

1 person liked this
Log inor Register

to rate and post a comment

(136)
(14)
(4)

Lin Dong

re: CTe artefact image

April 12, 2013 09:48

Reload the software,if not OK, your Detector or DAS may have problem.

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: CTe artefact image

April 12, 2013 09:57

You have likely made a mistake in the software load. Also, the fact that you can cal and it is okay for an hour likely means that you did not properly align the Xray tube when it was changed.

Log inor Register

to rate and post a comment

(56)
(15)

imeng

re: CTe artefact image

April 14, 2013 10:48

Hi,
Did you make all cal include Qcal and all were passed ok? I mean after the replacement of the new tube.
There is two parts in your problem, one could be solved by reloading the software, but regarding the arfifact which you said it appears after one hour of work, I guess this could be related to heat displacment, so I guess you have to pass the Qcal which will make exposures in both cases cols tube and hot tube. If I am in your place, I will start with tube re-alignment before re-install the software.

Stay well.

Log inor Register

to rate and post a comment

(56)
(15)

imeng

re: CTe artefact image

April 14, 2013 10:48

Sorry, I forgot to ask, is the artifact appears in all examinations or just head CT Scans?
regards,

Log inor Register

to rate and post a comment

(28)
(1)

Ronald

re: CTe artefact image

May 31, 2013 12:28

look likes if the tube become hot there is without artifact, try to heat up the tube before positioning X Ray tube adjustment, maybe

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)