Jairo Manzaneda

cte dual

July 07, 2015 06:16

i have this ct system with follow errors:
Colimator Request Timer Timeup.
11-10F0-03 TGP(MP) did not receive any reply from OGP in response to APERTURE CMD.
11-1010-03 TGP(MP) did not receive any reply from OGP in response to SysConfig CMD.
20-0014-02 OGP did not receive any reply from JEDI in response to SysConfig CMD nor JEDI
Capability Request CMD.
Some body have any idea about, i will appreciate for any techncial help.

1 person liked this
Log inor Register

to rate and post a comment

(29)
(11)
(16)

Vijay Sharma

re: cte dual

July 09, 2015 01:42

Hi, it seems there is no communication between tgp and rotating side of gantry which includes collimator , ogp and zedi.
1. Pls check slipring power 115 v Ac.
2. Pls check/clean slipring .
3. Check if ogp is booted properly and loop led is blinking.

Best Regards,

1 person liked this