There is a failure after scanning. There is no scouts after scanning. And the new patient do not creating in the scanning frame. System write errors via DASM DIAG procedure in Service Menu: result: Host computer did not detect DASM (scsi ID 3) at last boot up. What it can be?
during DASM diag. the system analyzes the following: hinv, scsistat, showdasm and rsp. Under scsistat, it checks the SCSI Bus, connectors and SCSI ID. First make sure that your DASM is powered and being run. Check on the led status of the DASM on the peripheral box in the OC. If you have a service manual, then check that you have the right led status for all most especially image transfer on the DASM-VDB. WHAT IS THE DASM START-UP ERROR VALUE? I think you have lost your DASM as they are not powered up, so check the connectors.
Hello, Craig! Is there the DASM based on the Linux PC 865 that named Recon? The Recon connected to the Host via Ethernet. The Host PC has two SATA HDD connected via USB to SATA adapter. Is DASM some interface between many devices (DAS, Host, Recon, printers, SCSI-bus)?
the DASM VDB (analog interface) and the DASM LCAM (digital interface) are used for filming purposes. CAN YOU PRINT?
With linux 865, the DASM via the Bit3 Backplane card to the SCSI. the DASM on this host is for 'Laser camera interface' and not recon.
the RECON engine runs through NPRIF (Interface), NPRM (Master) and NPRS (slave). they also are connected to the Bit3 Backplane Card.
Since you have problems with DASM and RECON, check your backplane card. you may need replacement.
Dmitry Dederer
CT/e problem via scanning
May 17, 2012 09:34
There is a failure after scanning. There is no scouts after scanning. And the new patient do not creating in the scanning frame. System write errors via DASM DIAG procedure in Service Menu: result: Host computer did not detect DASM (scsi ID 3) at last boot up. What it can be?
to rate and post a comment
ADEMETAN OLUWASEGUN
re: CT/e problem via scanning
May 18, 2012 10:28
what is the make of the ct machine
to rate and post a comment
S.DHINESH RAJ
re: CT/e problem via scanning
May 18, 2012 09:27
dear,
plz cheak das fuses and console dasifn fuse, and all
to rate and post a comment
CRAIG WAMBURA
re: CT/e problem via scanning
May 19, 2012 10:23
during DASM diag. the system analyzes the following: hinv, scsistat, showdasm and rsp. Under scsistat, it checks the SCSI Bus, connectors and SCSI ID. First make sure that your DASM is powered and being run. Check on the led status of the DASM on the peripheral box in the OC. If you have a service manual, then check that you have the right led status for all most especially image transfer on the DASM-VDB. WHAT IS THE DASM START-UP ERROR VALUE? I think you have lost your DASM as they are not powered up, so check the connectors.
to rate and post a comment
Dmitry Dederer
re: re: CT/e problem via scanning
May 20, 2012 03:38
Hello, Craig! Is there the DASM based on the Linux PC 865 that named Recon? The Recon connected to the Host via Ethernet. The Host PC has two SATA HDD connected via USB to SATA adapter. Is DASM some interface between many devices (DAS, Host, Recon, printers, SCSI-bus)?
to rate and post a comment
CRAIG WAMBURA
re: re: re: CT/e problem via scanning
May 23, 2012 01:36
the DASM VDB (analog interface) and the DASM LCAM (digital interface) are used for filming purposes. CAN YOU PRINT?
With linux 865, the DASM via the Bit3 Backplane card to the SCSI. the DASM on this host is for 'Laser camera interface' and not recon.
the RECON engine runs through NPRIF (Interface), NPRM (Master) and NPRS (slave). they also are connected to the Bit3 Backplane Card.
Since you have problems with DASM and RECON, check your backplane card. you may need replacement.
to rate and post a comment
Dmitry Dederer
re: re: re: re: CT/e problem via scanning
May 23, 2012 04:16
Thanks a lot! We reinstalled the software on the new HDDs and reconfigured the Laser Camera (Kodak 8150) and the CT is now working.