CT MODEL : CT/e single slice
Xtream console .
Linux Computer Type .
THE PROBLEM
Main Proble :
I have reloaded OS Software R./ V 4.5.1 CTT ) , and also reloaded the Application Software ( V/R 3.12 ) .
When I click on START UP to start the system , a window appears ( please see attached picture .)
Also , if I want to load Patch , I click on the SOFTWARE LOAD on the upper right side of the screen , and insert the CD and press OK , the System
says the NO CD IN THE DRIVE and asks me to check the CD ( it dose not detect the DVD WR DRIVE which is already configured as system device)
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
CT/e problem
April 21, 2014 11:46
CT MODEL : CT/e single slice
Xtream console .
Linux Computer Type .
THE PROBLEM
Main Proble :
I have reloaded OS Software R./ V 4.5.1 CTT ) , and also reloaded the Application Software ( V/R 3.12 ) .
When I click on START UP to start the system , a window appears ( please see attached picture .)
Also , if I want to load Patch , I click on the SOFTWARE LOAD on the upper right side of the screen , and insert the CD and press OK , the System
says the NO CD IN THE DRIVE and asks me to check the CD ( it dose not detect the DVD WR DRIVE which is already configured as system device)
Please advice .
to rate and post a comment
Carlos Duran
re: CT/e problem
April 22, 2014 10:04
Most probable, bad media. Then bad drive, or misconfiguration
to rate and post a comment
Galal
re: re: CT/e problem
April 22, 2014 05:52
is there any relation between DBPCI board missing and the problem of( power test fail ) when we start up the application ? enggalal.ha@gmail.com
to rate and post a comment
Carlos Duran
re: re: re: CT/e problem
April 23, 2014 09:13
Is a different scenario. If you get a fail test, is another issue, different to recognizing media
to rate and post a comment
Galal
re: re: re: re: CT/e problem
April 23, 2014 09:35
What it means when DBPCI is not found on booting ?
to rate and post a comment