CT/e single slice problem

June 08, 2014 08:48

After loading the application software , BDPCI board is not detected by the application Although it is new .

Please click on the link to see pictures. http://yadi.sk/d/8IdtUuodRhwDr

Appreciating your help .

1 person liked this
Log inor Register

to rate and post a comment

(40)
(6)
(1)

Santosh Kumar Tiwari

re: CT/e single slice problem

June 09, 2014 09:12

Where r u Located.

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: CT/e single slice problem

June 09, 2014 06:51

Yemen

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: CT/e single slice problem

June 09, 2014 09:12

Either your software is not loaded properly or you have not loaded all of the patches.

Log inor Register

to rate and post a comment

(67)
(7)

Galal

re: re: CT/e single slice problem

June 09, 2014 06:51

Regarding the patches , I could not install them because the system dose not detect the DVD Drive and window shows up to see NO CD ROM FOUND " CHECK CD ROM DRIVE '
Note : only DVD RAM drive is installed on machine and the softwares were installed via it.

Log inor Register

to rate and post a comment

(73)
(8)

Kim Schofield

re: re: re: CT/e single slice problem

June 10, 2014 09:22

That is your problem. Figure out what happened with the CD drive

Log inor Register

to rate and post a comment

(28)
(4)

MOI

re: CT/e single slice problem

June 09, 2014 04:06

SO THIS ERROR IS IN TEST NPR ? LAST TIME I HAVE THIS ERROR? I CHANGE DBBPI BOARD BUT IS NOT THIS BOARD FAILD. I CHANGE PCI BOARD PCIBACKPLANE AND IS OK. THE SYSTEM WAS AUTO SHUTDOWN BEFORE. I HOPE THIS CAN HELP YOU