Regarding CT/e CT Scanner- GE Make, what is the roll of the front panel board during OS booting ?
The troubleshooting says that OS booting failure may be caused by :
Host PC Power feedback CN4 to front panel fails .
Please advice
If front panel has problem, it will show the error message. You can take off the front panel, make a short connnection from the cable, the unit can run without the front panel.
There is a big cable which connect the Gantry and FRONT PANEL,take this one out, and put into the service board in upper coner. Where are you located? my Email:lindong123@hotmail.com
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
CT/e single slice
December 16, 2013 04:25
Hi every one
Regarding CT/e CT Scanner- GE Make, what is the roll of the front panel board during OS booting ?
The troubleshooting says that OS booting failure may be caused by :
Host PC Power feedback CN4 to front panel fails .
Please advice
to rate and post a comment
Lin Dong
re: CT/e single slice
December 16, 2013 07:07
If front panel has problem, it will show the error message. You can take off the front panel, make a short connnection from the cable, the unit can run without the front panel.
to rate and post a comment
Galal
re: re: CT/e single slice
December 17, 2013 08:27
Thank you for reply.
The connection to panel has 4 lines , please tell me which lines to be short-connected .
to rate and post a comment
Lin Dong
re: re: re: CT/e single slice
December 17, 2013 01:26
There is a big cable which connect the Gantry and FRONT PANEL,take this one out, and put into the service board in upper coner. Where are you located? my Email:lindong123@hotmail.com
to rate and post a comment
R.K.S singh
re: CT/e single slice
December 17, 2013 08:28
please email me at rohitsingh4u@hotmail.com
to rate and post a comment