I have a problem with the GE cti hispeed ct scan. When i do a scan, during scanogram the image will appear and you can view it, but when i do the next series image will not come up and you canoot view it. can you help me solve this problem...suggestions and answers are really appreciated..thank you
yeah definitely the images will be deleted dont reload the software first regenerate the database after that go to next step but in both case your images will be wash
If the scan is being completed without any errors but the images are not appearing then the images are probably stuck in recon que (you can confirm that from recon manager)..... try changing your raw data disk and run reconfig program with applications shutdown.. it might solve your problem
how can i change the raw data disk?itry loading the application software but the result is still the same. the images in the next series does not appear. When i also see the error log it says that recon cannot initialize the IG. No more atemptsto to initialize the IG will be done without a restart. error code155325.Recon could not reset IG board error no.13
First you pull out this put and insert it again if problem is solved alright if not then press the reset on ig board and then let me know did you regenrate the database or not as i told you
in most cases reloading the software solves it,some times the raw data archives becomes corrupt in the hard drive, when you reload sft. it will partition again and discard if any bad sectors of the disk. thus, getting back to business.
Hope this helps.
Jorge Lopez.
Hi dear friend i am E engineer, the last novenber i instaled
a ct scanner General Electric Hispeed QXi and all work ok but in
april 2012 the system failed began to boot and the recon status no
changed a few menuts the system shutdown one and other time. then
i run a saw the error log and show a CUP error
i disassembly the RSU and cleaned all boards and connectors and the
system work good again but just one mount and the error began
again. since may 25 i can“t repair the system i think that raw
data in hard drive is corrupted i changed the data disk but
when i tried reload the operative system show the next: Unable
to execute Xio (0) Pci (13)scsi (2) cdrom (6) partition (8)
sash 64: media not loaded
cannot load Xio(0) Pci (13) scsi (2) cdrom (6) partition (8) sash 64
-- media not loaded.
Unable to continue; press enter to return to menu:
then i though that cd rom drive has damaged and replaced it but
the system have the same problem, somebody told me that the
mother board in octane has damaged I changed the octane completly
and the ct no change boot OC but when charge apps shut down again
because datect a CUP error i tried reload the opertive system
but I having the same leyend:
Unable to execute Xio (0) Pci (13)scsi (2) cdrom (6) partition (8)
sash 64: media not loaded
cannot load Xio(0) Pci (13) scsi (2) cdrom (6) partition (8) sash 64
-- media not loaded.
Unable to continue; press enter to return to menu:
i
could you help me to resolve this problem ?
I made test_high_speed and show the next:
testing the high-speed disk by creting /raw_data/test_file.1.
Unable to open file /raw_data/test_file.1. aborting
Error # 2949129 errno = 0x2d0009
value = 1=0x1
tried reconfig scanDisk but is not possible the result is picture 2
One call to 440-263-6575 and your move will be covered.
We repair & store trailers. We build to your custom needs.
JDI Mobile Medical Manufacturing. Inc. We have your back.
Marivic Mendoza
ct/i hispeed ct scan
January 24, 2012 08:29
Good day,
I have a problem with the GE cti hispeed ct scan. When i do a scan, during scanogram the image will appear and you can view it, but when i do the next series image will not come up and you canoot view it. can you help me solve this problem...suggestions and answers are really appreciated..thank you
to rate and post a comment
John Cline
Re: ct/i hispeed ct scan
January 25, 2012 07:26
Try to reload the software. That fixes a lot of problems
to rate and post a comment
Marivic Mendoza
Re: ct/i hispeed ct scan
January 26, 2012 12:44
is there is another way to solve this problem without reloading the software. if i load the software the images that are store might be deleted.
to rate and post a comment
farhan zeb
Re: ct/i hispeed ct scan
January 26, 2012 11:58
yeah definitely the images will be deleted dont reload the software first regenerate the database after that go to next step but in both case your images will be wash
to rate and post a comment
Marivic Mendoza
Re: ct/i hispeed ct scan
January 26, 2012 12:44
when i run the system integration command the reconstruction hardware(IG) failed
to rate and post a comment
RICHIEK SHARMA
Re: ct/i hispeed ct scan
January 26, 2012 11:58
Dear Mr Mendoza
If the scan is being completed without any errors but the images are not appearing then the images are probably stuck in recon que (you can confirm that from recon manager)..... try changing your raw data disk and run reconfig program with applications shutdown.. it might solve your problem
to rate and post a comment
Marivic Mendoza
Re: ct/i hispeed ct scan
January 27, 2012 09:40
how can i change the raw data disk?itry loading the application software but the result is still the same. the images in the next series does not appear. When i also see the error log it says that recon cannot initialize the IG. No more atemptsto to initialize the IG will be done without a restart. error code155325.Recon could not reset IG board error no.13
to rate and post a comment
farhan zeb
Re: ct/i hispeed ct scan
January 28, 2012 01:51
Change the IG board and then try again
to rate and post a comment
Marivic Mendoza
Re: ct/i hispeed ct scan
January 29, 2012 11:41
i dont have available IG board.is there other way to fix it without changing the IG board.tnx
to rate and post a comment
farhan zeb
Re: ct/i hispeed ct scan
January 30, 2012 09:07
First you pull out this put and insert it again if problem is solved alright if not then press the reset on ig board and then let me know did you regenrate the database or not as i told you
to rate and post a comment
Marivic Mendoza
Re: ct/i hispeed ct scan
January 30, 2012 08:51
i already reset the ig board and press the reset button but still the same problem.how can regenerate the database?
to rate and post a comment
farhan zeb
Re: ct/i hispeed ct scan
January 31, 2012 08:25
what is your email address
to rate and post a comment
Marivic Mendoza
Re: ct/i hispeed ct scan
January 31, 2012 11:51
marivic_harmonix@yahoo.com
to rate and post a comment
SungHo Kang
Re: ct/i hispeed ct scan
February 20, 2012 10:21
did u fix its problem or still?
Pls let us know.
I can help you.
to rate and post a comment
Diego Barloqui
Re: ct/i hispeed ct scan
February 24, 2012 06:05
I have the same problem.. could you help me??
one of the raw data disks has broken and I need to replace it whitout reload software...
Thankyou in advance,..
to rate and post a comment
Diego Barloqui
Re: ct/i hispeed ct scan
February 24, 2012 11:37
I forgot that I'm working on Hispeed Nx/i (linux console)... thanks again!
to rate and post a comment
JORGE LUIS LOPEZ
Re: ct/i hispeed ct scan
January 26, 2012 11:16
in most cases reloading the software solves it,some times the raw data archives becomes corrupt in the hard drive, when you reload sft. it will partition again and discard if any bad sectors of the disk. thus, getting back to business.
Hope this helps.
Jorge Lopez.
to rate and post a comment
manuel valdez
re: Re: ct/i hispeed ct scan
September 21, 2012 04:26
Hi dear friend i am E engineer, the last novenber i instaled
a ct scanner General Electric Hispeed QXi and all work ok but in
april 2012 the system failed began to boot and the recon status no
changed a few menuts the system shutdown one and other time. then
i run a saw the error log and show a CUP error
i disassembly the RSU and cleaned all boards and connectors and the
system work good again but just one mount and the error began
again. since may 25 i can“t repair the system i think that raw
data in hard drive is corrupted i changed the data disk but
when i tried reload the operative system show the next: Unable
to execute Xio (0) Pci (13)scsi (2) cdrom (6) partition (8)
sash 64: media not loaded
cannot load Xio(0) Pci (13) scsi (2) cdrom (6) partition (8) sash 64
-- media not loaded.
Unable to continue; press enter to return to menu:
then i though that cd rom drive has damaged and replaced it but
the system have the same problem, somebody told me that the
mother board in octane has damaged I changed the octane completly
and the ct no change boot OC but when charge apps shut down again
because datect a CUP error i tried reload the opertive system
but I having the same leyend:
Unable to execute Xio (0) Pci (13)scsi (2) cdrom (6) partition (8)
sash 64: media not loaded
cannot load Xio(0) Pci (13) scsi (2) cdrom (6) partition (8) sash 64
-- media not loaded.
Unable to continue; press enter to return to menu:
i
could you help me to resolve this problem ?
I made test_high_speed and show the next:
testing the high-speed disk by creting /raw_data/test_file.1.
Unable to open file /raw_data/test_file.1. aborting
Error # 2949129 errno = 0x2d0009
value = 1=0x1
tried reconfig scanDisk but is not possible the result is picture 2
to rate and post a comment