Hello:
We have a problem with a GE CTi scanner. When trying to perform Warm up, or any other X Ray scan, we got the following messages:
Host: OBC
Generator Error: 183109
File: gen_abort.c 1.33
Function: X RAY GENERATION: KV Loop
Exception Level: Pri/Most Severe
KV Fault Detected: Anode Overcurret
Host: OBC Proc: Generator Error: 183296
Exception Level: Sec/Soft
Selected Technic: 80 KV, 50 mA, 0% Mod, Backup Timer 2160
Var mA: (Index: Start 0, Skip 0, Last 0) Last cmd 50 mA
Monitoring: (Measured 0.0 mA vs Ref 50 mA)
0 Total Samples, Tube Spit Ctr 0
I will appreciatte any help in dealing with this problem. Thanks.
Elias
Elias Valdes
CTi problem, error
December 05, 2007 07:28
Hello:
We have a problem with a GE CTi scanner. When trying to perform Warm up, or any other X Ray scan, we got the following messages:
Host: OBC
Generator Error: 183109
File: gen_abort.c 1.33
Function: X RAY GENERATION: KV Loop
Exception Level: Pri/Most Severe
KV Fault Detected: Anode Overcurret
Host: OBC Proc: Generator Error: 183296
Exception Level: Sec/Soft
Selected Technic: 80 KV, 50 mA, 0% Mod, Backup Timer 2160
Var mA: (Index: Start 0, Skip 0, Last 0) Last cmd 50 mA
Monitoring: (Measured 0.0 mA vs Ref 50 mA)
0 Total Samples, Tube Spit Ctr 0
I will appreciatte any help in dealing with this problem. Thanks.
Elias
to rate and post a comment
Ralph Childs
Re: CTi problem, error
December 06, 2007 12:19
Elias,
Save time... Call your local Xray tube supplier.
Scooter
scanner1@eim.ae
to rate and post a comment
Elliot Marcus
Re: CTi problem, error
December 07, 2007 12:08
I agree with Scooter.... Get a tube on order and replace it
to rate and post a comment
Elias Valdes
Re: CTi problem, error
December 19, 2007 09:58
Thank you for the help