Cutera Coolglide Error List

June 02, 2018 01:10

Hello, I live in Ireland and my Xeo gives me error code is 339. Can you assist me with this, even if you can help with how to fix book because there are no cutera technicians around here. Thanks

Log inor Register

to rate and post a comment

(14)
(2)

Peds

re: Cutera Coolglide Error List

June 03, 2018 04:04

Error 339- Flash HP detector mismatch.

1 person liked this
Log inor Register

to rate and post a comment

(14)
(2)

Peds

re: Cutera Coolglide Error List

June 03, 2018 04:04

Error 339- Flash HP detector mismatch.

Log inor Register

to rate and post a comment

(14)
(2)

Peds

re: Cutera Coolglide Error List

June 03, 2018 04:04

Flash HP detector mismatch.

Log inor Register

to rate and post a comment

(1)
(1)

Riels-Laser

re: Cutera Coolglide Error List

June 03, 2018 04:06

Are you able to ship the machine?. the problem seems to possibly be calibration

1 person liked this
Log inor Register

to rate and post a comment

(2)

mike smith

re: re: Cutera Coolglide Error List

June 04, 2018 11:08

Hi and thanks for the reply, if I post to you what will it cost to fix the problem.

Log inor Register

to rate and post a comment

(14)
(2)

Peds

re: re: re: Cutera Coolglide Error List

June 04, 2018 12:25

you need a cutera service manual, service programs and a laptop to communicate to the Coolglide machine in order to fix the calibration issue.

Log inor Register

to rate and post a comment

(13)

Parshenkov Nikolay

re: re: re: re: Cutera Coolglide Error List

June 05, 2018 09:42

Hello, could someone help me to solve some problems with cutera ce220xeo? Cutera Coolglide Interconnection Board burnt, and i cant identify some elements to change them. Concretly i cant identify u20.
Also would be greate to have service manual for this machine.

Log inor Register

to rate and post a comment

(1)

(1)

Zsacari Holland

re: re: re: Cutera Coolglide Error List

June 04, 2018 04:00

If all the machine needs is calibration then the labor fee would be $1100. If parts are needed then that would be extra. Give us a call at 6469329672 and we can better assist you and give you a better estimate.