DAS Control problem

(1)

(2)

Johnnie Owen

DAS Control problem

December 15, 2004 01:58

I have a DAS problem with a TCT-600. When I do a profile of the data from a scan on air the bell curve has a sawtooth waveform. The axial sacns have hard rings.

916 776 1836
jowen@teal.net

Log inor Register

to rate and post a comment

(57)
(1)
(15)

Barbaros Bulent Temiz

Re: DAS Control problem

December 25, 2004 05:50

dear johnnie,
Did you check the Sinogram of the image?If you see some vertical lines ,we can consider that problem would be from integrator or Adc.You should try also offset check with air than check the data of channels.
You can send me some picture data and we can try to help you.We have all boards for spare if you want.
blent temiz.
bbtek@ada.net.tr