Dexxum T Radius Total error.

November 26, 2015 10:00

Hello,

I have a problem with a Dexxum T, the Z score for the Radius Total in the forearm scanning is always around 10-12 although the T score is normal.

In the Radius 33, the T and Z scores are normal. And the rest of examinations for the hip and spine are ok.

I changed the detector and also did the re-calibration of the device a couple of times.

Does anyone has any suggestion that could help me fix the problem?


Regards,

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Dexxum T Radius Total error.

November 27, 2015 03:46

have you verified the Brush function is enabled (see user manual)I had this problem once.
I wish i could help you more but I do not have a service manual for this unit.

Log inor Register

to rate and post a comment

(14)
(4)

Albert Victor Stegaru

re: re: Dexxum T Radius Total error.

December 02, 2015 07:20

The Brush function is enabled, if I use the values change but they are still big on the total radius...
The system is working fine except this problem.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Dexxum T Radius Total error.

December 02, 2015 11:38

do you have the service manual for this system?

Log inor Register

to rate and post a comment

(14)
(4)

Albert Victor Stegaru

re: re: re: re: Dexxum T Radius Total error.

December 02, 2015 03:57

Yes i have it but is not complete o don't have the wire diagrams..

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: Dexxum T Radius Total error.

December 02, 2015 05:56

it is more than what I have. If it is PDF i would appreciate a copy please.
**redacted**