3712 at least one of me tubes programmed has not been adapted
lhis error is displayed always after switching on or RESET of CPU
if the adaptation has not yet been carried oul for a I I of the tubes (filaments)
• if error is not reselable: shor.drcuit in tube housing temperature contact
line EZX37 i EZX3a to ground?
We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com
Milos Krstic
DIagnost 92 error
December 02, 2014 10:17
What is meaning of error code 3712 on Philips Diagnost 92?
I'm thankful for any explanation.
Regards
to rate and post a comment
Lenny Place
re: DIagnost 92 error
December 03, 2014 08:33
3712 at least one of me tubes programmed has not been adapted
lhis error is displayed always after switching on or RESET of CPU
if the adaptation has not yet been carried oul for a I I of the tubes (filaments)
• if error is not reselable: shor.drcuit in tube housing temperature contact
line EZX37 i EZX3a to ground?
to rate and post a comment