Discovery W Error 349

Tags: Bone Densitometer, HOLOGIC

(78)
(11)
(1)

Joseph Hughes

Discovery W Error 349

September 05, 2012 05:32

I am having an error happen when there is a patient on the table. The error is #349. I do not have documentatin telling me what this problem is, and techsupport said to check the dicom que, then reset the unit. This worked for aoubt 2 patients, but now is a problem again. Now tech support says to fix the problem before using the systme on any more patients.
Any one know what i can do?

Thanks for your help,
Joseph

1 person liked this
Log inor Register

to rate and post a comment

(37)
(7)
(3)

Albert J. GARRY

re: Discovery W Error 349

September 05, 2012 07:09

Hello Joseph,

Contact me at aj.garry@gmail.com

In your email, let me know where you are located and send a number in case we are in the same time zone. I'm in central time zone.

Also, when yuo get the error, click on the "details" tab and let me know what other data is diplayed. With that, I will be able to better assist you.

AJ

Log inor Register

to rate and post a comment

(95)
(7)
(2)

Rick Ebling

re: Discovery W Error 349

September 06, 2012 10:26

It sounds like you have an IT problem. Maybe a slow line and things are timing out.

Log inor Register

to rate and post a comment

AL Papazian

re: Discovery W Error 349

September 06, 2012 06:00

Dear Joseph,
Please provide your e-mail address so we can help you.

Thanks

Log inor Register

to rate and post a comment

(78)
(11)
(1)

Joseph Hughes

re: Discovery W Error 349

September 06, 2012 06:38

It seems that our PACS cooridinator was doing some database restructuring, thus causing some down time or "congested" time on our PACS. I guess this modality is not as forgiving as some others are. Thanks everyone for your help!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)