Upon booting We are getting errors in IDPROM like IDPROM checksum error and
Invalid Format code in IDPROM.
we tried to download software but it failed at the application phase.
from past experience when I see these errors it is normally due to the generator having corrupt calibration data, I normally go into the service mode and clear the checksum then recalibrate
This issue came from the NVRAM battery is gone. Then,You need to replace the NVRAM on SUN Ultra10 or SUN Blade 150. (depend on your system) and after that you need to config the ID on the new NVRAM correspond to your SUN computer on Senographe 2000D system.
you have problem with the nvram of the system.We can fix it but you have to sent us the SUN computer to us. Sent E-MAIL for details at service@alfatech.gr
If you can contact us direct we can assist you on this issue without having to send your computer to us. email me please debi.breland@digitecmedical.com
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Jamal Bazzi
DMR 2000D
February 09, 2015 04:27
Upon booting We are getting errors in IDPROM like IDPROM checksum error and
Invalid Format code in IDPROM.
we tried to download software but it failed at the application phase.
to rate and post a comment
Lenny Place
re: DMR 2000D
February 10, 2015 10:57
from past experience when I see these errors it is normally due to the generator having corrupt calibration data, I normally go into the service mode and clear the checksum then recalibrate
to rate and post a comment
Natthakit Prongchai
re: DMR 2000D
February 10, 2015 10:57
This issue came from the NVRAM battery is gone. Then,You need to replace the NVRAM on SUN Ultra10 or SUN Blade 150. (depend on your system) and after that you need to config the ID on the new NVRAM correspond to your SUN computer on Senographe 2000D system.
to rate and post a comment
Jamal Bazzi
re: re: DMR 2000D
February 11, 2015 09:33
how to configure the ID on the new NVRAM
to rate and post a comment
CHRIS NALMPANTIS
re: DMR 2000D
February 10, 2015 11:00
you have problem with the nvram of the system.We can fix it but you have to sent us the SUN computer to us. Sent E-MAIL for details at service@alfatech.gr
to rate and post a comment
Charlie Britt
re: DMR 2000D
February 10, 2015 11:00
Hi, you have to replace the NVRAM in the sun computer.
to rate and post a comment
Deborah Breland
re: DMR 2000D
February 11, 2015 03:43
If you can contact us direct we can assist you on this issue without having to send your computer to us. email me please debi.breland@digitecmedical.com
to rate and post a comment