DMR PLUS ANODE ERROR

Tags: Mammo

Posted by an Anonymous User

DMR PLUS ANODE ERROR

May 31, 2016 10:24

Hi all, I have the E03 Rotot Failure 017/003, we have change:
1. Anode starter board
2. Generator command board
3. Generator interface board
4. Xray Tube
We also swapped/exchanged gantry and generator cpu boards and their programmed memories.
We measured 20ohms on each rotor conector.
250vdc
27vdc
5.15vdc
+12vdc
-12 vdc
240Ac
Rotor/tube cable unbroken...
Any help? Please Im crazy

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: DMR PLUS ANODE ERROR

May 31, 2016 06:54

did you try the generator command board 400PL1?

Log inor Register

to rate and post a comment

(22)

Douglas Varela

re: re: DMR PLUS ANODE ERROR

June 01, 2016 10:17

Yes it was changed already!

Log inor Register

to rate and post a comment

(36)
(4)

Ethan MacIntosh

re: re: DMR PLUS ANODE ERROR

June 01, 2016 11:01

Lenny is right. It's the 400PL1 that detects the 3phase from the tube stator currents.

Log inor Register

to rate and post a comment

(22)

Douglas Varela

re: re: re: DMR PLUS ANODE ERROR

June 01, 2016 12:24

Is there a way I can troubleshoot before changing it agai?
Voltages are +5.15+ 11.89 -12.14
From anode starter board I measure 4amps at each stator terminal...there is a high voltage at the begining and then it goes to 4amps and increases very slowly to 4.2 4.5..

Log inor Register

to rate and post a comment

(36)
(4)

Ethan MacIntosh

re: re: re: re: DMR PLUS ANODE ERROR

June 01, 2016 02:41

Not sure what the signal at TP44 (400PL1) should look like.

Log inor Register

to rate and post a comment

(36)
(4)

Ethan MacIntosh

re: re: re: re: re: DMR PLUS ANODE ERROR

June 03, 2016 02:08

With 5V at TP44, LED DS5 on 400PL1 should be on when trigger 1 is pressed which is Rotation Presence (TP 32) on 400PL2.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: DMR PLUS ANODE ERROR

June 01, 2016 02:41

yes did you get the manual I emailed earlier if so go to page 5-11 (pdf page 171 of 283 and follow the troubleshooting card TSG005

Log inor Register

to rate and post a comment

(86)
(5)

SHIJU KARAKKOOTTIL

re: DMR PLUS ANODE ERROR

June 02, 2016 09:48

Please let me know the following things.
1.CPU Board Version
2.D29 Diode Availability in 200 PL 1

Log inor Register

to rate and post a comment

(86)
(5)

SHIJU KARAKKOOTTIL

re: DMR PLUS ANODE ERROR

June 03, 2016 10:23

Have you got solution, if not please mail me on shiju@tecsila.com

Log inor Register

to rate and post a comment

(22)

Douglas Varela

re: re: DMR PLUS ANODE ERROR

June 03, 2016 12:59

I have 2325550-RV2 V6.22 cpu board.
I did not find D29 on 200pl1...it goes til D18.
Signal at TP44 is 4.75vdc when I press the first triger. On service the and without autotest the tube rotates at 6000rpm and 9000rpm...