DPX-BRAVO ROLLOVER

Tags: Bone Densitometer, GE

Posted by an Anonymous User

DPX-BRAVO ROLLOVER

July 20, 2013 07:56

When try to do something inmediatlly send AGS THRESHOLD ROLLOVER error,this system was stocked and I just installed the x ray tube, could the tube do this problem? It is posible adjust through service? I will need cal too, to do it?

thank you in advance

Log inor Register

to rate and post a comment

CHRIS NALMPANTIS

re: DPX-BRAVO ROLLOVER

July 22, 2013 10:15

you have to log in for service and then you go at signal monitor , first go at the settings and disable the ags rollover then start the signal monitor and try to move the colimator so that you take the best signal after that set max and you are done

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

re: re: DPX-BRAVO ROLLOVER

July 23, 2013 08:29

Thank you Chris, I did, but send same error

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: DPX-BRAVO ROLLOVER

July 22, 2013 10:20

You nee adjust the threshold, if you need asistence contact at semmedical@gmail.com

Log inor Register

to rate and post a comment

(75)
(17)
(7)

Ben Williams

re: DPX-BRAVO ROLLOVER

July 22, 2013 06:13

The system needs to be peaked and then aligned. I can walk you through that if you like. Ben: 608-514-2413

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

re: re: DPX-BRAVO ROLLOVER

July 23, 2013 08:29

thank you Ben, I contact you

Log inor Register