When try to do something inmediatlly send AGS THRESHOLD ROLLOVER error,this system was stocked and I just installed the x ray tube, could the tube do this problem? It is posible adjust through service? I will need cal too, to do it?
you have to log in for service and then you go at signal monitor , first go at the settings and disable the ags rollover then start the signal monitor and try to move the colimator so that you take the best signal after that set max and you are done
Sell your scrap MRI magnets! Top prices, free shipping, and fast pickups. Don't let dust-collecting, idle assets go to waste. Contact: ryan.swiderski@cyclicmaterials.earth or call 437-826-1621
DPX-BRAVO ROLLOVER
July 20, 2013 07:56
When try to do something inmediatlly send AGS THRESHOLD ROLLOVER error,this system was stocked and I just installed the x ray tube, could the tube do this problem? It is posible adjust through service? I will need cal too, to do it?
thank you in advance
to rate and post a comment
CHRIS NALMPANTIS
re: DPX-BRAVO ROLLOVER
July 22, 2013 10:15
you have to log in for service and then you go at signal monitor , first go at the settings and disable the ags rollover then start the signal monitor and try to move the colimator so that you take the best signal after that set max and you are done
to rate and post a comment
eduardo medina
re: re: DPX-BRAVO ROLLOVER
July 23, 2013 08:29
Thank you Chris, I did, but send same error
to rate and post a comment
Manuel Casanova
re: DPX-BRAVO ROLLOVER
July 22, 2013 10:20
You nee adjust the threshold, if you need asistence contact at semmedical@gmail.com
to rate and post a comment
Ben Williams
re: DPX-BRAVO ROLLOVER
July 22, 2013 06:13
The system needs to be peaked and then aligned. I can walk you through that if you like. Ben: 608-514-2413
to rate and post a comment
eduardo medina
re: re: DPX-BRAVO ROLLOVER
July 23, 2013 08:29
thank you Ben, I contact you
to rate and post a comment