Over 500 Total Lots Up For Auction at One Location - NJ 06/15

DPX-IQ ramping problem

(77)
(1)
(4)

eduardo medina

DPX-IQ ramping problem

June 16, 2008 04:26

I have a dpx-IQ with ramping problem,the error log tell me a 80(+HVPS ou of range)I try with other pair of HVPS that ramping on other system, but here simply not work.Could be another thing?.

Any suggestions?

Eduardo

1 person liked this
Log inor Register

to rate and post a comment

(39)
(1)
(2)

Emerson Freire

Re: DPX-IQ ramping problem

June 16, 2008 06:18

You could have a tube problem...

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

Re: DPX-IQ ramping problem

June 17, 2008 09:20

I am trying only with HVPS and HV cables out of receptacules of HVPS,and only try to apply KV, but the system not ramping,But repeat in other system that HVPS ramping, but cannot get good graph.

Log inor Register

to rate and post a comment

(8)

wilson cavalcante

re: Re: DPX-IQ ramping problem

May 15, 2014 08:04

I have a problem on a DPX IQ presenting error HVPS. Which way can I go to repair? if you can help me send email to wcscavalcante@uol.com.br
Abs
Wilson.

Log inor Register

to rate and post a comment

(139)
(11)
(2)

ALEXANDER PILAFTSIS

Re: DPX-IQ ramping problem

June 17, 2008 03:20

Eduardo,
For the first case it might be any part of the HV system. X-ray tube, HVPS both, control cable, HV cables or even max and oink board. On the second case probably you have not connect properly the HVPS. I can not imagine anything else or a fuse has been blown.
Is your software version 4.7e? If yes please send me the error log to specify the problem.

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

Re: DPX-IQ ramping problem

June 17, 2008 09:23

tomorrow will send you error log. But I am only applying KV on static counter,with cables out of receptacules of PS. and not ramp.

thank you

Log inor Register

to rate and post a comment

(77)
(1)
(4)

eduardo medina

Re: DPX-IQ ramping problem

June 18, 2008 10:45

Alexander i dont know to where send error log

Thank you

Log inor Register

to rate and post a comment

(33)
(3)
(12)

Abel Cardona

Re: DPX-IQ ramping problem

June 20, 2008 10:54

maybe you can try ramping without x ray tube.
unplug x ray cables from the x ray tube and try ramping (be careful), you mus see arcing high voltage plugs. if not, you doesn´t have h.v. tension. if i can help you my email is siemeds@hotmail.com

Log inor Register

to rate and post a comment

(8)

wilson cavalcante

re: DPX-IQ ramping problem

May 15, 2014 08:04

I have a problem on a DPX IQ presenting error HVPS. Which way can I go to repair? if you can help me send email to wcscavalcante@uol.com.br
Abs
Wilson.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Don’t Compromise on Safety or Speed

From skilled GE depot repairs to patient-ready accessories, Southeastern Biomedical specializes in GE to keep your facility safe, efficient, and cost-effective. 888-310-7322.

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)