normal items have oxyshure-technology, you can see text at booting. In the config you can change to masimo. some mainboards need therefore a change of a jumper in the near of the multimed-connector. Than you have to connect sthe masimo-pod and masimo sensor. Also some items have nellcor oxy-max (label right und, at front). then you have to connect Oxy-max-Pod and this sensor. Lokk in the biomed-service-menu, code: 4712, bedside-config, if there is nellcor or masimo. For changing you need a code from draeger. good luck.
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
simone sottile
Drager InfinityDelta
February 28, 2020 10:21
Hello ,
Drager Infinity Delta spo2 sensor incompatible message
Help me ..please !!!!
to rate and post a comment
juan pablo amado
re: Drager InfinityDelta
March 03, 2020 10:50
Here is service manual, it is equal to drager
http://www.frankshospitalworkshop.com/equipment/documents/ecg/service_manuals/Siemens_SC7000-9000XL_-_Service_Manual.pdf
to rate and post a comment
Igor69
re: Drager InfinityDelta
March 04, 2020 09:55
normal items have oxyshure-technology, you can see text at booting. In the config you can change to masimo. some mainboards need therefore a change of a jumper in the near of the multimed-connector. Than you have to connect sthe masimo-pod and masimo sensor. Also some items have nellcor oxy-max (label right und, at front). then you have to connect Oxy-max-Pod and this sensor. Lokk in the biomed-service-menu, code: 4712, bedside-config, if there is nellcor or masimo. For changing you need a code from draeger. good luck.
to rate and post a comment