I have a drystar 3000 with an error code 6c466. I have asked everyone i know who works on the drystar and ive looked in the manual and cant seem to find what this error code means. Any help will be appreciated thank you.
COS sw module not be installed
reason:
a) hardware error (cos board not inserted prorly; board broken)
b) config error: file cos.prg not present in disk or corrupted
c)system error: tasks not succeed
cure: verify flat cables, ck Cos board and vme rack
ck release installed,verify hdd
ck config of the engine
ck cpu and operating system components; reinstall sw
use another add
Error Code --> 6c466
E_T14_COS_INSTALL,
Meaning : COS software module could not be installed.
Check returned error for details.
Reason : Depending on the error mentioned as the final
error (indicating the error-family) we can
distinguish 3 types of error :
a. Hardware error ...
The COS-board is not inserted properly inside
the VME-rack. The board might also be broken.
b. Configuration error ...
The file is either not present on the
disk or is corrupted.
c. System error ...
Some of the system functions (such as creating
tasks) did not succeed.
Cure : a. Check the COS-board and the VME-rack itself.
Verify the flatcables. Replace the board as
a test.
b. Check the release installed. Verify the
harddisk functionality. Use the tools for doing
this.
c. Check the basic configuration of the engine.
Check th
The COS1 Board Revision D can be differentiated from the previous version, revision C, by a modified firmware and modified hardware.
If a COS1 Board Revision C is replaced by a board with revision D, the following restrictions apply always depending on the software in the machine:
Type 5361/999 (up to SN 1068)
These machines were shipped with the COS1 Board Revision C and the Software Release 1.32. Prior to the installation of a COS1 Board Revision D, a Software Release 1.47 or higher
must be installed. This software version is available via MED-CSO upon request.
Type 5361/999 (up to SN 1269 except 1262, 1267)
These machines were shipped with a COS1 Board Revision C and Software Release 1.35, 1.38 or 1.39. The installation of Software Release 1.42, 1.43, or 1.45 will result in error message 6C466 60248.
Remedy: In case of this problem the COS1 Board Revision D must be installed.
Note: Prior to the installation of the software release 1.4X ( 60248.
Remedy: In case of this problem the COS1 Board Revision D must be installed.
Note: Prior to the installation of the software release 1.4X (
warren meyer
DRYSTAR 3000 Error Code 6c466
August 21, 2015 11:33
I have a drystar 3000 with an error code 6c466. I have asked everyone i know who works on the drystar and ive looked in the manual and cant seem to find what this error code means. Any help will be appreciated thank you.
to rate and post a comment
juan pablo amado
re: DRYSTAR 3000 Error Code 6c466
August 21, 2015 04:06
COS sw module not be installed
reason:
a) hardware error (cos board not inserted prorly; board broken)
b) config error: file cos.prg not present in disk or corrupted
c)system error: tasks not succeed
cure: verify flat cables, ck Cos board and vme rack
ck release installed,verify hdd
ck config of the engine
ck cpu and operating system components; reinstall sw
use another add
to rate and post a comment
Lenny Place
re: DRYSTAR 3000 Error Code 6c466
August 21, 2015 04:06
Error Code --> 6c466
E_T14_COS_INSTALL,
Meaning : COS software module could not be installed.
Check returned error for details.
Reason : Depending on the error mentioned as the final
error (indicating the error-family) we can
distinguish 3 types of error :
a. Hardware error ...
The COS-board is not inserted properly inside
the VME-rack. The board might also be broken.
b. Configuration error ...
The file is either not present on the
disk or is corrupted.
c. System error ...
Some of the system functions (such as creating
tasks) did not succeed.
Cure : a. Check the COS-board and the VME-rack itself.
Verify the flatcables. Replace the board as
a test.
b. Check the release installed. Verify the
harddisk functionality. Use the tools for doing
this.
c. Check the basic configuration of the engine.
Check th
to rate and post a comment
kco
re: DRYSTAR 3000 Error Code 6c466
August 23, 2015 10:08
Modification of the COS1 Board in Revision D
Reason:
The COS1 Board Revision D can be differentiated from the previous version, revision C, by a modified firmware and modified hardware.
If a COS1 Board Revision C is replaced by a board with revision D, the following restrictions apply always depending on the software in the machine:
Type 5361/999 (up to SN 1068)
These machines were shipped with the COS1 Board Revision C and the Software Release 1.32. Prior to the installation of a COS1 Board Revision D, a Software Release 1.47 or higher
must be installed. This software version is available via MED-CSO upon request.
Type 5361/999 (up to SN 1269 except 1262, 1267)
These machines were shipped with a COS1 Board Revision C and Software Release 1.35, 1.38 or 1.39. The installation of Software Release 1.42, 1.43, or 1.45 will result in error message 6C466 60248.
Remedy: In case of this problem the COS1 Board Revision D must be installed.
Note: Prior to the installation of the software release 1.4X ( 60248.
Remedy: In case of this problem the COS1 Board Revision D must be installed.
Note: Prior to the installation of the software release 1.4X (