Drystar3K error 6C841 HELP!!!
April 06, 2010 07:08
Need help with Agfa Drystar 3000 that is giving error 6C841. It looks like the slit sensor S8 that measures drum rotation is bad. I tested the sensor and it is ok, evan IMOS sensor test said it is ok. I also tested PMC board and it looks ok.
IMOS reported the following condition:
Error reported by T14:
code(0x6c841->0x60248)
Any ideas? Is it a bad sensor or a bad board?
to rate and post a comment
yuan bo
Re: Drystar3K error 6C841 HELP!!!
April 11, 2010 12:41
please contact me, i can help you. obnauy@163.com
to rate and post a comment
Bao Vo Quoc
Re: Drystar3K error 6C841 HELP!!!
June 10, 2010 10:16
Error Code --> 6c844
E_T14_MDM_SCAN_HOME,
Meaning : At the moment of beginning an MDM-scan
operation, the sensor read-out states that
the wagon stays at the HOME-position.
The motor is supposed to be driven but leaving
the HOME-position is not detected.
Reason : a. The sensor monitoring the MDM's HOME-position
is defective.
b. The motor does not move, even if it is
attempted to be moved (no DC-voltage present ?).
Cure : a. Check the MDM-home sensor (s5) functionality.
b. Check the power connections to motor M6 (MDM).
c. Check the motor itself.
d. Check the MDM-board (CAN-based).
e. Reset of system after correction is requested.
Vo Quoc Bao
Field Service Engineer