DT 60 Matematical error help
October 31, 2007 01:16
I can“t calibrate, it send a matematical error in calibrate process
TKS fro help me.
to rate and post a comment
Gbenga Adeola
Re: DT 60 Matematical error help
November 01, 2007 11:54
Ernest,
you have problem with the memory, change the battery on the main pc board, carry out the white and black reference adjustments and do your calibration.
from Gbenga Adeola