E10 senograph error

(31)
(2)

catur suharinto

E10 senograph error

January 26, 2008 01:01

dear everyone
I have a problem with senograph 800 (E10 heating error track 1 022/002). the error is appear intermittenly (after 5th exposure). I have checked the thermostat(tube housing) and fillament, everything is OK.
can somebody help me, thank you
sorry for my english

Log inor Register

to rate and post a comment

Umberto Sciunnach

Re: E10 senograph error

January 28, 2008 05:34

Put outside the connector from tube.
Check the waret.
regards uto

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: E10 senograph error

January 28, 2008 11:53

thank you for respons, but I'm sorry what is warret.

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: E10 senograph error

January 28, 2008 11:57

Sounds like the computer dosen't think that the filament is there. It could be a poor connection in the High voltage probe(candle stick)either at the tube or tansformer. Other possiblilities include the feedback cables from the transformer to the 400PL1 PL2 boards. Worst case, it could be the tube, after it heats up a little after 3-4 exposures the filament may open.

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: E10 senograph error

January 28, 2008 09:34

Thank you for response. I checked the fillament using ohmeter, both of large or small track is still connected (no problem).

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: E10 senograph error

January 29, 2008 09:49

Contact me through e-mail, shayward@sodexhoctm.com
I can send you a few more things to try / check.

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: E10 senograph error

January 30, 2008 07:09

CATUR:

IN EXPERIENCES THE ERROR APPEARS WHEN THE ARC IN TUBE DURING IN XRAY EXPOSURE. THE PROBLEM IS IN TRACK 1 ONLY, PROBE WITH THE SMALL FILLAMENT, TRACK 2, IF YOU NO HAVE THE PROBLEM WITH THESE.

FOR YOUR PROBLEMS ARE ARC IN TUBE, THE SYSTEM 800T BLOCKED THE EQUIP WHEN THE 5 ARC IN TUBE APPEARS., YOU OFF THE EQUIP AND THE PROBLEM IS SOLVED.

CHANGE THE TUBE.

BEST REGARDS.

MANUEL

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: E10 senograph error

February 09, 2008 02:08

thak you for respons
are you sure abaut change the tube, because you know that tube is an expensive part.

Log inor Register

to rate and post a comment

(244)
(19)

Staate Hayward

Re: E10 senograph error

February 12, 2008 03:42

I would tend to agree about the tube being the problem. If both large and small filaments arc at high KV and/or longer exposures then the tube is OLD or Gassy. There is a tube warm-up procedure in the service manual for the gassy problem, that may help if the tube isn't OLD, (Cal. job card 10). If the problem is only associated with one focal spot then the tube is even more suspect. Sorry I couldn't get back to this sooner, I tried to send a PDF file but it was to lagre (even win-ziped) to e-mail to you.

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: E10 senograph error

February 15, 2008 12:52

may you upload that file to 4shared or easy share.?

Log inor Register

to rate and post a comment

(31)
(2)

catur suharinto

Re: E10 senograph error

February 19, 2008 10:23

dear mr manuel
I'm sorry, you are wrong abaut the tube. the main problem is aux board 300 pl11
thank you for response

Log inor Register

to rate and post a comment