E-cam E25

July 27, 2017 02:14

Hello, please what must to do in this case in e-cam, a motion control has accursed E25 .. Camera reconfiguration.? Thank you

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: E-cam E25

July 28, 2017 09:00

EE# 25
RC_UNRECOVERABLE

A serious detector reconfiguration system error has occurred that the firmware cannot recover from on its own.

Correct the mechanical or sensor problem and enter a valid angle through the Diagnostics User Interface.Possible FRUs:
Reconfiguration Latch, Head 2 @ 180 Flag, Rings locking Mechanism, Rings locked Sensor.

1 person liked this
Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: E-cam E25

July 28, 2017 09:01

Hello. Could be.
But could me other thigs, can you send me more details? I can help you. I have many parts and intructions. Send me all details on daniloabdo@gmail.com

We can figure out this...

1 person liked this
Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: E-cam E25

July 28, 2017 09:01

Hello. Could be.
But could me other thigs, can you send me more details? I can help you. I have many parts and intructions. Send me all details on daniloabdo@gmail.com

We can figure out this...

1 person liked this
Log inor Register

to rate and post a comment

(18)
(5)
(1)

Jorge Barrios

re: E-cam E25

July 28, 2017 09:03

I think you have any problem in rotation movement, you have to check potenciometer in big motor there is behind and check flats, one flat is in motor support and another flat when you open back cover on the top you can see shutter and next to shutter there is another flat. Check those elements work propertly and calibrate rotation movements.
Also, sometimes you can have problem with AMP board, check every green led on AMPs are lighting.
Good luck

1 person liked this
Log inor Register

to rate and post a comment

(7)

Sik

re: re: E-cam E25

July 30, 2017 08:43

Thank you very much For your message I Will be Try

Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: E-cam E25

July 28, 2017 03:41

Did you check all voltages? 24v /12v? All works on diagnostic mode?

1 person liked this
Log inor Register

to rate and post a comment

(7)

Sik

re: re: E-cam E25

July 30, 2017 08:43

Yes all work in diagnostic mode.

Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: re: re: E-cam E25

July 31, 2017 10:18

Can you show us here the Log error? On service menu ( service key needed )

1 person liked this
Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: re: re: E-cam E25

July 31, 2017 10:18

Can you show us here the Log error? On service menu ( service key needed )

1 person liked this
Log inor Register

to rate and post a comment

(7)

Sik

re: re: re: re: E-cam E25

July 31, 2017 10:37

Hello, thank's for your message, so The error is gone but after I have "The gantry is not calibrated", I do not know what I need to do ?

Log inor Register

to rate and post a comment

(7)

Sik

re: re: E-cam E25

July 31, 2017 10:18

Yes all work in only in manuel mode

Log inor Register

to rate and post a comment

(20)
(8)

Danilo Abdo

re: re: re: E-cam E25

July 31, 2017 05:37

Try reconver gantry calibrarion on service mode. Is necessary a service key to do it, but is a faster and safe way. On service mode, you can calibrate all motions individualy too if necessary.

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

Probo Medical | We Buy, Repair, and Sell New and Used Ultrasound Equipment

We provide sales, service, rental, and repair for ultrasound systems, probes, and C-arms. We have the largest inventory in the world. Contact us about our special offers on certified refurbished equipment! Call: 317-759-9210 Email: sales@probomedical.com.

stats

Upcoming Auction Events

Cbrx2sales
TIMED
Jun 05, 2026 (157 lots)