Easy Diagnost Imaging system

Tags: Radiology - R/F, PHILIPS

(3)
(1)

Fred Jaramillo

Easy Diagnost Imaging system

August 07, 2012 02:32

Is anyone experiencing premature camera failure in Philips Easy Diagnost Imaging Systems. Our XTV 21 Camera Head 4522 163 27945 seems to be going out every year. We are not under a service agreement and we feel that it is not power related. I am curious if others are experiencing this issue on or off of a service agreement.

Log inor Register

to rate and post a comment

(1)

Donn Cooper

re: Easy Diagnost Imaging system

August 08, 2012 10:12

What is "going out" on the camera? There are several known problems on XTV19-21 cameras. Let me know the exact problem and I will let you know if it is one of these problems.

Log inor Register

to rate and post a comment

(3)
(1)

Fred Jaramillo

re: re: Easy Diagnost Imaging system

August 08, 2012 03:08

Both systems experienced the same basic symptoms, for a few weeks prior to a total failure we would get TV/II (TV/Image Intensifier) not ready, error log would say camera control self test failed, reboot system and system would boot up and work all day.

Log inor Register

to rate and post a comment