Easy Diagnost with Super 80 generator image problem.

Tags: Radiology - R/F

(181)
(27)

Tibor Ivanszky

Easy Diagnost with Super 80 generator image problem.

November 12, 2017 02:17

Dear friends,
I am looking for an expert on an Easy Diagnost system.
The problem is that at the beginning of the fluoroscopy the image is good but within seconds the kV is increasing in two steps by few kilovolts and the image becomes bad.
It is ADC or ABS or image chain problem, i do not know how to point the troubled area.
Pls, write to tibor@ivanszky.com

Log inor Register

to rate and post a comment

(3)

Ernest Morgan

re: Easy Diagnost with Super 80 generator image problem.

November 18, 2017 03:49

Try using kv lock so technique does not change and see if image is stable. If image deterioration continues it is the camera. What xtv11 camera is it

Log inor Register

to rate and post a comment