Over 500 Total Lots Up For Auction at One Location - NJ 06/15

efilm configuration with agfa printers

Tags: PACS / DR / CR / Digital

(4)

Adeyemi TAIWO

efilm configuration with agfa printers

February 04, 2019 09:22

Please how can I configure my DRYSTAR AXYS with my efilm software for X-ray and CT printing. Thanks

Log inor Register

to rate and post a comment

(41)
(2)
(1)

Henry Lopez

re: efilm configuration with agfa printers

February 04, 2019 01:18

You find your Dicom info for all modalities (AETitle,IP address,port#) then you add them into your efilm station

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: re: efilm configuration with agfa printers

February 06, 2019 08:48

More to it than that

Log inor Register

to rate and post a comment

(33)

Awais

re: efilm configuration with agfa printers

February 04, 2019 01:19

Contact me
Awaiskhan271992@gmail.com

Log inor Register

to rate and post a comment

(332)
(69)

Gabriel Cool

re: efilm configuration with agfa printers

February 04, 2019 01:19

Which CT and Xray? You want print from efilm ?

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: efilm configuration with agfa printers

February 05, 2019 08:36

You need the ae title and ip. I'll have to look at my configuration table for my efilm settings to give you further information.

Log inor Register

to rate and post a comment

(315)
(43)
(1)

Ronald Swapp

re: re: efilm configuration with agfa printers

February 06, 2019 08:41

There is more to setting up the printer in efilm than just adding the ip address and ae title unlike what other individuals are telling you.

Get the IP address and AE title from the printer, port will be 104. Go to keypad of the printer and show info, from there IP address, AE and port will be accessible.

Open efilm and go to Edit, then open the Dicom Printers tab.
Bottom section under Device create an alias or Description. Call it AXYS
AE Title: enter the AE of the printer, might be AXYS but you can find it from the key pad under show info.

IP Address: enter IP address of printer
Port: enter 104
Format: Standerd \1,1 (this means it will be a 1x1 format.
Priority: medium
Medium: Depending on what your film is if it's DT2B set it as Blue
If it's DT2C set it as clear.
Copies: set as 1
Film Orientation: set as Portrait
Film Size: depends on what you have in the drawers, if you want to only print on 14x17 set 14INx17IN
Film Destination: set as Processor
Resolution: set as 300
Magnification: set as Cubic
Smoothing: set as Medium
Trim: set as Yes
Densities section
Min: set as 0
Border: Black
Max: 300 (this is the d-max of the printer)
Empty: 300
Click Add
Then click: Verify
If it's successful it will verify, if not double check your ip address and ae title.
If you need to make changes to the printer highlight it from your list make the necessary changes and click update.

Log inor Register

to rate and post a comment

(410)
(57)

CRAIG WAMBURA

re: efilm configuration with agfa printers

February 05, 2019 08:45

Contact me via WhatsApp +256782702243 and we get this done quickly.