1.This machine connect wiht CT scan?, Manufacture? Model?.
2. When you power on. " Ready" displayed on pad panel?.
3. Check signal cable between printer & pad panel,check temperature on Laser unit in service mode.
4.Do you have " Tera term" software?. This software can help you connect with printer by RS232 and loading error log.
Online auction of modern high-end laboratory and biopharma research equipment from the recent bankruptcy of Spexis AG in Allschwil near Basel, Switzerland. Start bidding today! closes on June 4. info@dome-auctions.com..+31(0)20 808 2880
Vu Toan
ELK-4000 Print Problem
August 14, 2008 12:57
pls help me
The Unit can not print test, calib but it has not any error on control panel
Thanks & best regards
Vu Toan
to rate and post a comment
Pham PhuQuoc
Re: ELK-4000 Print Problem
September 04, 2008 12:05
Hi Toan.
Please supply more details trouble informations.
I can help you cover this trouble.
1.This machine connect wiht CT scan?, Manufacture? Model?.
2. When you power on. " Ready" displayed on pad panel?.
3. Check signal cable between printer & pad panel,check temperature on Laser unit in service mode.
4.Do you have " Tera term" software?. This software can help you connect with printer by RS232 and loading error log.
Please give me above information.
Thanks!
Pham Phu Quoc
to rate and post a comment
Ronald Liem
Re: ELK-4000 Print Problem
April 08, 2010 03:46
I need service manual or diagram for ELK 3000/4000.
Do you have?, please send me to: ronaldliem@yahoo.com
Thanks and best regards.