Elscint selectsp

November 20, 2009 09:50

Being in good operating order, suddenly the numeric display, starts flashing and the scanner hangs. I have to boot off of nuevo.Que all can be?

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Elscint selectsp

November 26, 2009 07:33

Hi,
have you got any informations in errorlog? Maybe there are something with the display connections under the front cover or on the RMS&COMM card.
Does system change to status SAFETY? Can you hear that rotor stops at the same time?

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Elscint selectsp

November 26, 2009 07:33

Hi,
have you got any informations in errorlog? Maybe there are something with the display connections under the front cover or on the RMS&COMM card.
Does system change to status SAFETY? Can you hear that rotor stops at the same time?

Log inor Register

to rate and post a comment

(110)
(10)
(5)

Wojciech Selwesiuk

Re: Elscint selectsp

December 03, 2009 02:47

Hi again,
If your GANTRY goes to hang and you have to restart it it is possible that your RMS&COMM board or MCU module has a communication or software problem. I had the situation similar to your in the past (I don't remember now which module was replaced - it was 2 years ago - but one of them)

Log inor Register

to rate and post a comment

(24)
(1)

dino mantegazza

Re: Elscint selectsp

December 04, 2009 08:21

Hi thanks for replying, now having this problem is random, in terms of rms module already had problems.
Dino Greetings

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Re: Elscint selectsp

August 12, 2015 07:37

You will have information on how to configure a new dicom node?

Log inor Register

to rate and post a comment

(432)
(83)
(3)

juan pablo amado

re: Re: Elscint selectsp

August 12, 2015 07:37

You will have information on how to configure a new dicom node?