Over 150 Missouri Auctions End Tomorrow 06/05 - Bid Now

epilight simmer error

Tags: Cosmetic

(4)
(2)

peter martin

epilight simmer error

November 19, 2014 09:45

Any ideas as to why our beloved epilight has shut down with simmer error code 104 . Connection = short. We had some capacitor problems, replaced offending capacitor. Now one week later we have this problem

1 person liked this
Log inor Register

to rate and post a comment

(15)
(6)

Jose Nelson Gouveia Vogado

re: epilight simmer error

December 16, 2014 08:36

reemplace treatment head

1 person liked this
Log inor Register

to rate and post a comment

(4)
(2)

peter martin

re: re: epilight simmer error

June 13, 2016 10:18

That worked thank you.
Now I have a new problem. the machine has been shutting down with error code 94 - short . Any ideas?

1 person liked this
Log inor Register

to rate and post a comment

(15)
(6)

Jose Nelson Gouveia Vogado

re: re: re: epilight simmer error

June 13, 2016 12:24

Hi Martin, the code error #94-95 is the following: Replace Switch Module. maybe, is the IGBT is short cicuit? revise betwen Emissor colector, Good is open ciscuit, bad is short circuit

1 person liked this
Log inor Register

to rate and post a comment

(4)
(2)

peter martin

re: re: re: re: epilight simmer error

June 13, 2016 06:10

I think you are on the money again! It was, shutting down intermittently prior. After giving it a good clean inside , it would fire up.

Log inor Register