We have Picker PQ 2000 CT Scanner.The machine is working and also the machine is warming up regularly.But we dont take any images.
Can you please give us information regarding this problem.
1004 error is ofteh associated with a problem in collimator assembly. One servo is nor working and you need to identify which is the one that produce the error.
Using MANEXP program you can ove each by separate and try to make a single shoot. Use 80KV, 30mA 1second, speed 0 and modify one by one: filter, compensator, collimator and field.
You also can check connections at RRCB just at top of signal brushes.
Ercan TUTUNCUOGLU
Eror Code:1003(Picker PQ 2000 CT Scanner)
July 10, 2009 11:41
We have Picker PQ 2000 CT Scanner.The machine is working and also the machine is warming up regularly.But we dont take any images.
Can you please give us information regarding this problem.
Best Regards,
Ercan Tutuncuoglu
to rate and post a comment
Mazher saeed
re: Eror Code:1003(Picker PQ 2000 CT Scanner)
May 23, 2013 05:52
XRAY ERROR,System in interlock
when i want to make warm up
i checked all switches please guide me that how can i solve this problme
yhnx
to rate and post a comment
Domingo Perez
re: Eror Code:1003(Picker PQ 2000 CT Scanner)
May 24, 2013 10:15
LOW_SPEED_ROTOR_SOFTWARE_ERROR
Possible BAD STARTER, RCB, CABLES, FUSES, XSC, EXCHANGER
to rate and post a comment
Mazher saeed
re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)
May 30, 2013 03:01
hi
i,m experiencing the problem in my IQ Picker CT scanner
that
first of All
1)when i take warmup machine gets process and warms up
2)when i take pilot machine takes pilots also
some time i take pilot no pilot perfom and couch move silently out )
3)when i take images slices then it does not take and show an error 2012
please guide and help me thanks
to rate and post a comment
Mazher saeed
re: Eror Code:1003(Picker PQ 2000 CT Scanner)
June 26, 2013 02:13
i,m getting ERR 1004 when want to make warmup
some time it take warm up and on images also get err 1004
SOME time also it is Ok
but gantry move CLOCK vise during scanning
please guide me
thanks
to rate and post a comment
Jose Morillo
re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)
June 26, 2013 11:43
1004 error is ofteh associated with a problem in collimator assembly. One servo is nor working and you need to identify which is the one that produce the error.
Using MANEXP program you can ove each by separate and try to make a single shoot. Use 80KV, 30mA 1second, speed 0 and modify one by one: filter, compensator, collimator and field.
You also can check connections at RRCB just at top of signal brushes.
Feel free for contact me for more assistance
to rate and post a comment
Mazher saeed
re: re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)
July 02, 2013 05:06
Thanks please also guide me
WHEN I ON GANTRY IN SERVICE MODE
Gantry move v v slowly
some time does not move but i move it by hand
WHEN SYSTEM IS IN NORMAL MODE:
when i want to warm up IQ premier Extra
Gantry does Not warmup but some time move Wrong position duriNg warmup
when i take pilot it takes pilot
But when i want to make slice Images
after 2 or 3 slice gantry move very fast(OVER SPEED)in correct position some time go in wrong position back
I changed SERVO BOARD on LEFT UPPER side in GANTRY
i Chekced PMDC and GPUC are both ok also
to rate and post a comment
Mazher saeed
re: re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)
July 05, 2013 05:56
Thanks please also guide me
WHEN I ON GANTRY IN SERVICE MODE
Gantry move v v slowly
some time does not move but i move it by hand
WHEN SYSTEM IS IN NORMAL MODE:
when i want to warm up IQ premier Extra
Gantry does Not warmup but some time move Wrong position duriNg warmup
when i take pilot it takes pilot
But when i want to make slice Images
after 2 or 3 slice gantry move very fast(OVER SPEED)in correct position some time go in wrong position back
I changed SERVO BOARD on LEFT UPPER side in GANTRY
i Chekced PMDC and GPUC are both ok also
to rate and post a comment
Jose Morillo
re: re: re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)
July 06, 2013 12:44
zero detector board is bad.