Eror Code:1003(Picker PQ 2000 CT Scanner)

(12)

(1)

Ercan TUTUNCUOGLU

Eror Code:1003(Picker PQ 2000 CT Scanner)

July 10, 2009 11:41

We have Picker PQ 2000 CT Scanner.The machine is working and also the machine is warming up regularly.But we dont take any images.
Can you please give us information regarding this problem.

Best Regards,
Ercan Tutuncuoglu

Log inor Register

to rate and post a comment

(36)

Mazher saeed

re: Eror Code:1003(Picker PQ 2000 CT Scanner)

May 23, 2013 05:52

XRAY ERROR,System in interlock

when i want to make warm up
i checked all switches please guide me that how can i solve this problme

yhnx

Log inor Register

to rate and post a comment

(25)
(8)

Domingo Perez

re: Eror Code:1003(Picker PQ 2000 CT Scanner)

May 24, 2013 10:15

LOW_SPEED_ROTOR_SOFTWARE_ERROR
Possible BAD STARTER, RCB, CABLES, FUSES, XSC, EXCHANGER

Log inor Register

to rate and post a comment

(36)

Mazher saeed

re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)

May 30, 2013 03:01

hi

i,m experiencing the problem in my IQ Picker CT scanner
that
first of All

1)when i take warmup machine gets process and warms up

2)when i take pilot machine takes pilots also
some time i take pilot no pilot perfom and couch move silently out )

3)when i take images slices then it does not take and show an error 2012

please guide and help me thanks

Log inor Register

to rate and post a comment

(36)

Mazher saeed

re: Eror Code:1003(Picker PQ 2000 CT Scanner)

June 26, 2013 02:13

i,m getting ERR 1004 when want to make warmup
some time it take warm up and on images also get err 1004

SOME time also it is Ok
but gantry move CLOCK vise during scanning
please guide me

thanks

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)

June 26, 2013 11:43

1004 error is ofteh associated with a problem in collimator assembly. One servo is nor working and you need to identify which is the one that produce the error.

Using MANEXP program you can ove each by separate and try to make a single shoot. Use 80KV, 30mA 1second, speed 0 and modify one by one: filter, compensator, collimator and field.
You also can check connections at RRCB just at top of signal brushes.

Feel free for contact me for more assistance

Log inor Register

to rate and post a comment

(36)

Mazher saeed

re: re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)

July 02, 2013 05:06

Thanks please also guide me

WHEN I ON GANTRY IN SERVICE MODE

Gantry move v v slowly
some time does not move but i move it by hand

WHEN SYSTEM IS IN NORMAL MODE:
when i want to warm up IQ premier Extra
Gantry does Not warmup but some time move Wrong position duriNg warmup

when i take pilot it takes pilot
But when i want to make slice Images

after 2 or 3 slice gantry move very fast(OVER SPEED)in correct position some time go in wrong position back


I changed SERVO BOARD on LEFT UPPER side in GANTRY

i Chekced PMDC and GPUC are both ok also

Log inor Register

to rate and post a comment

(36)

Mazher saeed

re: re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)

July 05, 2013 05:56

Thanks please also guide me

WHEN I ON GANTRY IN SERVICE MODE

Gantry move v v slowly
some time does not move but i move it by hand

WHEN SYSTEM IS IN NORMAL MODE:
when i want to warm up IQ premier Extra
Gantry does Not warmup but some time move Wrong position duriNg warmup

when i take pilot it takes pilot
But when i want to make slice Images

after 2 or 3 slice gantry move very fast(OVER SPEED)in correct position some time go in wrong position back


I changed SERVO BOARD on LEFT UPPER side in GANTRY

i Chekced PMDC and GPUC are both ok also

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

re: re: re: re: Eror Code:1003(Picker PQ 2000 CT Scanner)

July 06, 2013 12:44

zero detector board is bad.