Your Email:
Francisco Junior
Erro code C-48 ERBE vio 300 D
May 30, 2020 02:24
Hello, would someone please tell me the error shown on the Erbe Vio 300 D display? "Activation was canceled due to a failure. If the fault occurs, inform the technical assistance service C-48" Thank you!
to rate and post a comment
Sergey Radionov
re: Erro code C-48 ERBE vio 300 D
June 01, 2020 11:04
Measurement value for phase too small with ON. Check generator (TP generator). Check sensors. Recognizing module: C = CPU + Sensors
Fateh Fateh
re: re: Erro code C-48 ERBE vio 300 D
March 23, 2023 10:17
Hi please can you help me for error B-E
Zsolt Torok
July 07, 2020 11:44
Message is normally caused by the Power Supply (30140-810), but the HF-Generator or CPU+Sens. can be also the trigger.
Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Francisco Junior
Erro code C-48 ERBE vio 300 D
May 30, 2020 02:24
Hello, would someone please tell me the error shown on the Erbe Vio 300 D display?
"Activation was canceled due to a failure. If the fault occurs, inform the technical assistance service C-48"
Thank you!
to rate and post a comment
Sergey Radionov
re: Erro code C-48 ERBE vio 300 D
June 01, 2020 11:04
Measurement value for phase too small with ON.
Check generator (TP generator).
Check sensors.
Recognizing module: C = CPU + Sensors
to rate and post a comment
Fateh Fateh
re: re: Erro code C-48 ERBE vio 300 D
March 23, 2023 10:17
Hi please can you help me for error B-E
to rate and post a comment
Zsolt Torok
re: Erro code C-48 ERBE vio 300 D
July 07, 2020 11:44
Message is normally caused by the Power Supply (30140-810), but the HF-Generator or CPU+Sens. can be also the trigger.
to rate and post a comment