Error 1 - Loss of the +15V dc
Check the U2 power supply and the cable connection at the U2.
I have this service manual
Best regards
Eduardo A. Tognon
BRAZIL
Thank you for sending me the manual , what can I do for you?.
I am going to look for the service manual HDI 3000.
If you need anything, don't hesitate to ask me. You can count with a friend in Bolivia.
Capacitor charge error
-Has the system not been used for a long period of time (higher capacitor leakage current)?
If the error occurs after switching the system off and back on again, configure the capacitor bank (refer to the "service program" chapter pages 4-2).
Are the dircharge resistors permanently connected?
Has the temperature sensors on D950 responded (D950 TP "TP")?
has the ovrevoltage in the C-bank been detected (D950 TP "VC")?
Is there a short on D950. The error appears only when the short circuit is continuous
Replace D950
Hi Henrique
Thanks for respond to me. This error 01 appears every time when I turn on the system.
I have been running the test 1 in service mode and the display changes number values going up and down. Are the capacitors charging and discharging.?
How much does it cost me the service manual ?
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Edgar Aranibar
Error 01 polymobil plus
November 21, 2008 07:09
I have Error 01 in a Polymobil plus siemens RX, could anyboby tell me what this means?
Also I need the service manuals.
Thank you
Edgar Aranibar
Bolivia
to rate and post a comment
Eduardo Tognon
Re: Error 01 polymobil plus
November 26, 2008 02:18
Error 1 - Loss of the +15V dc
Check the U2 power supply and the cable connection at the U2.
I have this service manual
Best regards
Eduardo A. Tognon
BRAZIL
to rate and post a comment
Edgar Aranibar
Re: Error 01 polymobil plus
November 29, 2008 09:47
Hi Eduardo
Thanks for respond to me, I check the voltages on U1 power supply and they are ok. But the U2 power supply provides +12Volts. Am I right or wrong?
Greetings
Edgar A. Quiroz
to rate and post a comment
Eduardo Tognon
Re: Error 01 polymobil plus
December 01, 2008 05:40
I send the service manual for your email
Best regards
Eduardo
to rate and post a comment
Edgar Aranibar
Re: Error 01 polymobil plus
December 01, 2008 09:35
Hi Eduardo
Thank you for sending me the manual , what can I do for you?.
I am going to look for the service manual HDI 3000.
If you need anything, don't hesitate to ask me. You can count with a friend in Bolivia.
Greetings
Edgar A. Quiroz
to rate and post a comment
Edgar Aranibar
Re: Error 01 polymobil plus
December 16, 2008 08:58
Hello Eduardo
I find the problem on my polymobil plus, there were a BUP318 and a Bussmann Semiconductor 20A,600V fuse broken. Do you have any of these componenents?
Grettings
Edgar Aranibar
to rate and post a comment
Carlos Villanueva
Re: Error 01 polymobil plus
June 20, 2009 12:33
Hi Eduardo and Edgar
I have a problem with the polymobil plus, it show an error 34 and if you can, please send me the service manual. I will be very grateful.
Regards from Venezuela,
Carlos Villanueva
to rate and post a comment
nga Nguyen
Re: Error 01 polymobil plus
April 07, 2012 10:51
can you send me 01 copy of service manual for siemens polymobil plus. thank you very much.
my email: phuongec@gmail.com
to rate and post a comment
Evandro Mariano da Silva
re: Re: Error 01 polymobil plus
May 20, 2014 10:24
Hi Edgar,
Could you send me this Service Manual?
email :evandro@biotecmedical.com.br
Thank you!
Evandro Mariano
to rate and post a comment
Evandro Mariano da Silva
re: Re: Error 01 polymobil plus
May 20, 2014 10:24
Hi Eduardo,
Could you send me this Service Manual?
Thank you!
Evandro Mariano
to rate and post a comment
Henrique Gomes
Re: Error 01 polymobil plus
November 27, 2008 06:56
Hello,
err 1 Polymobil Plus
Capacitor charge error
-Has the system not been used for a long period of time (higher capacitor leakage current)?
If the error occurs after switching the system off and back on again, configure the capacitor bank (refer to the "service program" chapter pages 4-2).
Are the dircharge resistors permanently connected?
Has the temperature sensors on D950 responded (D950 TP "TP")?
has the ovrevoltage in the C-bank been detected (D950 TP "VC")?
Is there a short on D950. The error appears only when the short circuit is continuous
Replace D950
I have the manual
Henrique Gomes
neutrinno@zipmail.com.br
to rate and post a comment
Edgar Aranibar
Re: Error 01 polymobil plus
November 29, 2008 09:45
Hi Henrique
Thanks for respond to me. This error 01 appears every time when I turn on the system.
I have been running the test 1 in service mode and the display changes number values going up and down. Are the capacitors charging and discharging.?
How much does it cost me the service manual ?
Greetings
Edgar A Quiroz
to rate and post a comment
Qaiser Anwar
Re: Hard copy of Siemens Polymobil Plus
November 19, 2009 02:14
Please send me the hard copy of manuals on my email (qanwer4all@yahoo.com).
Thanking u
to rate and post a comment