Over 150 Missouri Auctions End Today - Bid Now

ERROR 101 -10 IN SYTEC 2000i

(343)
(83)
(13)

Manuel Casanova

ERROR 101 -10 IN SYTEC 2000i

March 01, 2008 06:37

HI ALL:

I HAVE A PROBLEM IN SYTEC 2000i, WHEN MAKE A WARM UP PROCEDURES, THE ERROR CODE 101-10 APPEARS INMEDIATLY, I NEED THE ADVANCE SERVICE MANUAL, OR YOUR OPINION ABOUT THIS.

MANUEL

Log inor Register

to rate and post a comment

Ralph Childs

Re: ERROR 101 -10 IN SYTEC 2000i

March 01, 2008 11:03

Manuel,

Error 101-10 is an azimuth rotation error. Most probably your azimuth pot has either slipped or is defective. You will also need to scope the encoder output if the pot is okay.

Scooter
scanner1@eim.ae

Log inor Register

to rate and post a comment

(74)
(5)
(10)

Tayfun Canli

Re: ERROR 101 -10 IN SYTEC 2000i

March 06, 2008 10:11

Hi Manuel

Please perform Po,Pg adjustment. Refer to function check and adjustment book which must be located with system.