Hi thanks for help.Error 12 when was programin tube type filament apears Error 12,and nao error 12 apears allways and dont let me work, can you help
whit the trobleshutting Thanks
Er 12 Tube filament initializing not done
When first taken into use after tube head or generator processor replacement the tube filament
pre-heating values must be initialized. In the service mode this will be automatically done when
exposure switch is continuously pressed till operation stops. If this is not done, an attempt to
make a normal exposure will give this error indication.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Alfredo Castanos
error 12 PLANMECA PROLINE EC
May 13, 2016 09:15
Hi thanks for help.Error 12 when was programin tube type filament apears Error 12,and nao error 12 apears allways and dont let me work, can you help
whit the trobleshutting Thanks
comedica@hotmail.com
to rate and post a comment
Lenny Place
re: error 12 PLANMECA PROLINE EC
May 14, 2016 05:23
you have to run thru the initialization
Er 12 Tube filament initializing not done
When first taken into use after tube head or generator processor replacement the tube filament
pre-heating values must be initialized. In the service mode this will be automatically done when
exposure switch is continuously pressed till operation stops. If this is not done, an attempt to
make a normal exposure will give this error indication.
to rate and post a comment