ERROR 13 AIRSHIELD C-400
October 21, 2013 06:26
SOME ONE, CAN HELP ME PLEASE
I HAVE A AIRSHILED C-400 this show me error 13
Thanks a lot
to rate and post a comment
Milan Kubeck
re: ERROR 13 AIRSHIELD C-400
October 23, 2013 04:13
According to the service manual the "Heater control circuitry is defective". If you don't have a manual go to link below page 77 for the error codes. Looks like the board has test points so troubleshoot from there.
http://www.frankshospitalworkshop.com/equipment/documents/infant_incubators/service_manuals/Hill-Rom_Isolette_-_Repair_manual.pdf
to rate and post a comment
Francisco Jaramillo
re: re: ERROR 13 AIRSHIELD C-400
October 24, 2013 01:38
Milán let me see ur link THANKS YOU SO MUCH¡,
to rate and post a comment
Milan Kubeck
re: ERROR 13 AIRSHIELD C-400
October 23, 2013 04:14
According to the manual Error 13 is a "defective heater control circuitry defective". see page 13 on link below.
http://www.frankshospitalworkshop.com/equipment/documents/infant_incubators/service_manuals/Hill-Rom_Isolette_-_Repair_manual.pdf
Please excuse if this is a 2nd post
to rate and post a comment
Francisco Jaramillo
re: re: ERROR 13 AIRSHIELD C-400
October 24, 2013 01:38
THANKS X 1000