hello good day I had problems with x-ray kodak 2200 send me error message 13 I would like to know how to solve your problem and also if you could send me the service manual.
I hope you can help me and thank you very much for your attention
You have a broken lamp voltage divider.
By this you have incorrect measurement kV
You must dismantle the lamp and repair or replace the resistors in the voltage divider
in 99% this is the cause of error 13
I have a kodak 2200 intra oral Xray unit that is giving me the same error 13. My local technician has told me that this happened because this unit was not getting much use and that the tubehead has failed and is not worth repairing. Can you be of some help and tell me where to find this voltage divider resistors please?
orlando
error 13 kodak 2200
January 15, 2018 09:05
hello good day I had problems with x-ray kodak 2200 send me error message 13 I would like to know how to solve your problem and also if you could send me the service manual.
I hope you can help me and thank you very much for your attention
my mail is
orlando_1221@hotmail.com
to rate and post a comment
petroveech
re: error 13 kodak 2200
January 16, 2018 09:20
sent in the mail
to rate and post a comment
fortronik
re: error 13 kodak 2200
January 16, 2018 09:21
You have a broken lamp voltage divider.
By this you have incorrect measurement kV
You must dismantle the lamp and repair or replace the resistors in the voltage divider
in 99% this is the cause of error 13
to rate and post a comment
Etienne Cassar
re: re: error 13 kodak 2200
February 09, 2023 09:50
I have a kodak 2200 intra oral Xray unit that is giving me the same error 13. My local technician has told me that this happened because this unit was not getting much use and that the tubehead has failed and is not worth repairing. Can you be of some help and tell me where to find this voltage divider resistors please?