Over 150 Missouri Auctions End Today - Bid Now

Error 16 GE 600T

yojin ramones

Error 16 GE 600T

March 25, 2006 08:50

I have error 16 in Ge 600T when the unit is on, and I can not selec focus 0.3 only smal is work.
Could some body tell me what is the problem??

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

Re: Error 16 GE 600T

March 25, 2006 09:08

cause 1: The heater current board 400pl1, or heater power supply 400pl2 are not in position.
action 1: insert the board in slots.
cause 2:the large spot heater is not working properly
action 2: run the heater test job card lg004
cause 3 and just cause: the tube filament large is out,
Action 3:Replace the XRay tube.

bye

Log inor Register

to rate and post a comment

(9)

(1)

Joel Inacio

Re: Error 16 GE 600T

September 02, 2006 07:53

Insert a lamp (12v) between ground and small of hv cable. If the code error desapear the problem is in the tube.

Log inor Register

to rate and post a comment

Frank Beltre

Re: Error 16 GE 600T

February 05, 2007 10:02

Yojin,

I lost your email. Can you send it to me again.

Frank Beltre

Log inor Register

to rate and post a comment

stats