Over 150 Missouri Auctions End Today - Bid Now

ERROR 20 IN LORAD MIV

Tags: Mammo

(343)
(83)
(13)

Manuel Casanova

ERROR 20 IN LORAD MIV

December 12, 2013 05:10

ERROR 20 IN LORAD M IV, PLEASE ANY IDEAS??

I CHECK THE WIRES, AND CONNECTIONS, FUSES, BUT NOT WORK.

THANKS A LOT

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: ERROR 20 IN LORAD MIV

December 13, 2013 08:06

System Error 20 Rotor Control Open thermal switch or rotor fault Possible thermal overload. Allow 2
hours to cool. Check fuses

Log inor Register

to rate and post a comment

(15)

(1)

Curtis Wheelhouse

re: ERROR 20 IN LORAD MIV

December 13, 2013 08:09

Check tube thermal switch, F 21 on DC panel, 525v on rotor board, and rotor control EPROM. Aldo when you open the left side door the wires can come off of the large capacitor at the bottom of the door. It is the rotor starter cap. Good luck.

Log inor Register

to rate and post a comment

(15)

(1)

Curtis Wheelhouse

re: re: ERROR 20 IN LORAD MIV

December 13, 2013 04:16

Please update once you find the cause.

Log inor Register

to rate and post a comment

(1)

ramon moran

re: re: re: ERROR 20 IN LORAD MIV

August 29, 2022 10:54

Hi dear, I have the same error. I checked f21, the 500 volts led is on. I can't find the tube thermal switch.......can you tell where is located

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: ERROR 20 IN LORAD MIV

August 31, 2022 12:29

open the cover on the back side of the tube and check there.

did you ohm out the stators on thwe tube control board? to rule out a open wire

Log inor Register

to rate and post a comment

(343)
(83)
(13)

Manuel Casanova

re: re: ERROR 20 IN LORAD MIV

December 13, 2013 04:17

Curtis:

where is located the thermal switch?? I have 675 on the capacitor when xray press, the fuses is ok.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: ERROR 20 IN LORAD MIV

December 15, 2013 03:08

on the tube

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: re: re: ERROR 20 IN LORAD MIV

May 27, 2017 04:11

Hi Lenny, where on the tube is the thermal switch located, cause I see nothing attached on the tube. I appreciate your help.

Log inor Register

to rate and post a comment

(17)
(2)

antoninus andres

re: re: re: re: ERROR 20 IN LORAD MIV

May 30, 2017 10:13

hi lenny,
im also having this error 20 in lorad mIV.
what is the probable cause for this?

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: ERROR 20 IN LORAD MIV

May 30, 2017 02:04

This is a rotor issue I am sending you the schematics I hope it helps you

Log inor Register

to rate and post a comment

(17)
(2)

antoninus andres

re: re: re: re: re: re: ERROR 20 IN LORAD MIV

May 31, 2017 09:04

thanks alot,
this is my email: jamandres.ece@gmail.com

Log inor Register

to rate and post a comment

(6)
(2)

ARMANDO AGARIJO

re: re: re: re: re: re: ERROR 20 IN LORAD MIV

April 07, 2018 03:58

Hi lenny,
Please, I need the schematics of this equipment. My email is armandoagarijo@hotmail.com.

Regards

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: re: re: re: re: ERROR 20 IN LORAD MIV

April 10, 2018 08:51

okay i will send to you

1 person liked this
Log inor Register

to rate and post a comment

Engr Robert

re: re: ERROR 20 IN LORAD MIV

July 21, 2018 01:20

hello sir i have also err0r 20 in lorad M-IV
i check fuse 21 ok,Dc panel is ok, replace rotor control board but system is not ok . kindly help me
my email is biotechmurtaza214@gmail.com

Log inor Register

to rate and post a comment

(17)
(2)

antoninus andres

re: ERROR 20 IN LORAD MIV

May 30, 2017 10:13

hi Manuel, have you already found out the error 20 problem? i'm having this problem too..

Log inor Register

to rate and post a comment

(17)
(2)

antoninus andres

re: ERROR 20 IN LORAD MIV

May 30, 2017 10:13

have you already fix this problem?
im having this error too..

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: ERROR 20 IN LORAD MIV

July 23, 2017 04:45

Hi Antoninus, have you solved the issue already?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)