Hello Dears,
i have problem with Hologic Lord M-IV, When i command for exposure tube make noise and error display error 20.x rays does not emmitted
kindly tell me the solution.
i have replaced all pwb s of machine
The error 20 means problem with tube rotor or rotor board check the igbt , rotor relay on the rotor control board , check the start capacitor . And if is all ok so think that maybe the tube rotor coli some time are bad
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Engr Robert
Error 20 Lord M-IV
July 21, 2018 01:20
Hello Dears,
i have problem with Hologic Lord M-IV, When i command for exposure tube make noise and error display error 20.x rays does not emmitted
kindly tell me the solution.
i have replaced all pwb s of machine
to rate and post a comment
carlos enrique salazar gea
re: Error 20 Lord M-IV
July 22, 2018 02:08
System Error 20 Rotor Control.- Open thermal switch or rotor fault. Possible thermal overload. Allow 2
hours to cool. Check fuses.
to rate and post a comment
carlos enrique salazar gea
re: Error 20 Lord M-IV
July 22, 2018 02:08
You can download the service manual in internet
to rate and post a comment
Horacio Pérez
re: Error 20 Lord M-IV
July 22, 2018 02:10
Hello
Check for fuse F21 in DC panel
Check +525 volts on rotor board
and rotor control board EPROM
to rate and post a comment
Joel Grullon
re: Error 20 Lord M-IV
July 22, 2018 02:11
The error 20 means problem with tube rotor or rotor board check the igbt , rotor relay on the rotor control board , check the start capacitor . And if is all ok so think that maybe the tube rotor coli some time are bad
to rate and post a comment