Error 20 Lord M-IV

Tags: Mammo, HOLOGIC

Engr Robert

Error 20 Lord M-IV

July 21, 2018 01:20

Hello Dears,
i have problem with Hologic Lord M-IV, When i command for exposure tube make noise and error display error 20.x rays does not emmitted
kindly tell me the solution.
i have replaced all pwb s of machine

1 person liked this
Log inor Register

to rate and post a comment

(45)
(8)

carlos enrique salazar gea

re: Error 20 Lord M-IV

July 22, 2018 02:08

System Error 20 Rotor Control.- Open thermal switch or rotor fault. Possible thermal overload. Allow 2
hours to cool. Check fuses.

Log inor Register

to rate and post a comment

(45)
(8)

carlos enrique salazar gea

re: Error 20 Lord M-IV

July 22, 2018 02:08

You can download the service manual in internet

1 person liked this
Log inor Register

to rate and post a comment

(6)

Horacio Pérez

re: Error 20 Lord M-IV

July 22, 2018 02:10

Hello

Check for fuse F21 in DC panel
Check +525 volts on rotor board
and rotor control board EPROM

Log inor Register

to rate and post a comment

(102)
(13)

Joel Grullon

re: Error 20 Lord M-IV

July 22, 2018 02:11

The error 20 means problem with tube rotor or rotor board check the igbt , rotor relay on the rotor control board , check the start capacitor . And if is all ok so think that maybe the tube rotor coli some time are bad

Log inor Register

to rate and post a comment