I work with MIV Lorad, it start have a noise the tube and some time appear error 20, and after 30. well on day stop absolutely with error 20 and after error 30.
Check I saw in the rotor control board, the K2 burn, in pin 8 and 5, check the board change the K2 have the 500v and 90V, when use conetc with the tube, only turn on the mammo K2 start bourn, but when not connect the tube, the board is ok and the K2. Could be the tube bad?
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Yojin Ramones
error 20 MIV help
April 19, 2016 08:09
I work with MIV Lorad, it start have a noise the tube and some time appear error 20, and after 30. well on day stop absolutely with error 20 and after error 30.
Check I saw in the rotor control board, the K2 burn, in pin 8 and 5, check the board change the K2 have the 500v and 90V, when use conetc with the tube, only turn on the mammo K2 start bourn, but when not connect the tube, the board is ok and the K2. Could be the tube bad?
to rate and post a comment
Juancho
re: error 20 MIV help
April 22, 2016 09:32
Hi, Yojin
you have a trouble with Tube rotor or thermal switch. Send me your email
juancho
to rate and post a comment
Yojin Ramones
re: re: error 20 MIV help
April 22, 2016 04:27
thanks to all
it was the board damage, I change and work again
to rate and post a comment
antoninus andres
re: re: error 20 MIV help
September 30, 2016 10:19
hi sir, i have error 20 problem.. i would like to ask where is the thermal switch located? is it inside tyhe tube?
to rate and post a comment