Dear
I need help with Ge SIGNA-
The error are :
2225540
2225542
The system work fine with body coil, but with Head and other coil with central connector appear the error above.
Please let me know what do you suggest me
Thank you
depending on which version of Signa, likely candidates are the SSM (most likely SSM board is CPD) or Driver Module or RFSC; or the TR switch in the "Dog House"
I can repair you parts, I do not stock parts.
IF you call or email me I can give you the names of several very reliable GE parts sources. 720-340-4593, Bruce@medsystech.com
2225540
"Dynamic disable fault, Front of Coil short Dynamic Disable bias fault on front drivers. Possibly a short circuit load or a shorted diode in the dynamic disable switch boxes on FRONT of BRM."
2225542
"Dynamic disable fault, Rear of Coil short Dynamic Disable bias fault on back drivers. Possibly a short circuit load or a shorted diode in the dynamic disable switch boxes on BACK of BRM."
The problem maybe caused by: DD driver path.
You can disconnect the DD cable at the Penetration Panel, and measure the voltage of the the diode respectively to clarify the front or back DD switch boxes.
Giuseppe Castellano
ERROR 2225540 SIGNA
September 26, 2012 05:23
Dear
I need help with Ge SIGNA-
The error are :
2225540
2225542
The system work fine with body coil, but with Head and other coil with central connector appear the error above.
Please let me know what do you suggest me
Thank you
to rate and post a comment
Bruce Smith
re: ERROR 2225540 SIGNA
September 27, 2012 12:21
What else does the system tell you about the failure, what is the description of the fault?
to rate and post a comment
Giuseppe Castellano
re: re: ERROR 2225540 SIGNA
September 27, 2012 12:58
INDUCTIVE OR DIRECT DRIVE FAULT, OR BODY TR SWITCH FAULT DUE TO IMPEDENCE....
THANK YOU IN ADVANCED
to rate and post a comment
Bruce Smith
re: re: re: ERROR 2225540 SIGNA
September 27, 2012 01:56
depending on which version of Signa, likely candidates are the SSM (most likely SSM board is CPD) or Driver Module or RFSC; or the TR switch in the "Dog House"
to rate and post a comment
Giuseppe Castellano
re: re: re: re: ERROR 2225540 SIGNA
September 27, 2012 04:46
Dear
Thank you for your support.
Do you have the spare parts to test?
and let me know in which way we can collaborate?
I'll wait your answer
to rate and post a comment
Bruce Smith
re: re: re: re: re: ERROR 2225540 SIGNA
September 27, 2012 05:33
I can repair you parts, I do not stock parts.
IF you call or email me I can give you the names of several very reliable GE parts sources. 720-340-4593, Bruce@medsystech.com
to rate and post a comment
Marco Flores
re: re: ERROR 2225540 SIGNA
October 01, 2012 04:48
I agree with Bruce Smith: I have had this fault a couple of times and the bad board was CPD in SSM. We have the parts, but we are in Mexico
to rate and post a comment
Giuseppe Castellano
re: re: re: ERROR 2225540 SIGNA
October 01, 2012 05:51
Dear
Please give me your contact, and there'll be no problem with distance.
Thank you
to rate and post a comment
Marco Flores
re: re: re: re: ERROR 2225540 SIGNA
October 05, 2012 02:22
sorry for delay. ing_en_resonancia@yahoo.com.mx
to rate and post a comment
Dejun Li
re: ERROR 2225540 SIGNA
September 27, 2012 10:22
2225540
"Dynamic disable fault, Front of Coil short Dynamic Disable bias fault on front drivers. Possibly a short circuit load or a shorted diode in the dynamic disable switch boxes on FRONT of BRM."
2225542
"Dynamic disable fault, Rear of Coil short Dynamic Disable bias fault on back drivers. Possibly a short circuit load or a shorted diode in the dynamic disable switch boxes on BACK of BRM."
The problem maybe caused by: DD driver path.
You can disconnect the DD cable at the Penetration Panel, and measure the voltage of the the diode respectively to clarify the front or back DD switch boxes.
to rate and post a comment