This out of the manual
*Error 253 occurs when the system does not boot fully and displays the "system error
#253" message. This message is associated with a loss of serial communication.
Do you want a copy of the manual? reply to:
cbet_us@yahoo.com
Lightly used system available direct from the end user customer. MONARCH QUEST
Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.
Kenneth Calhoun
ERROR 253 OEC 9600
March 24, 2012 05:05
does anyone know what to do on a OEC9600 with a bootup error of 253
to rate and post a comment
Juan de la Cruz Martinez
Re: ERROR 253 OEC 9600
March 24, 2012 07:46
This message is associated with a loss of serial communication.
to rate and post a comment
Kenneth Calhoun
Re: ERROR 253 OEC 9600
March 25, 2012 12:30
Is this serial communication between the control panel board and the aux board??
to rate and post a comment
thai tan
Re: ERROR 253 OEC 9600
March 24, 2012 07:46
you should check contact on boards, cables. for me, you reassemble the most boards in the system to maintenance. good luck.
Thai
to rate and post a comment
Lenny Place
Re: ERROR 253 OEC 9600
March 25, 2012 11:07
this is most likely the interconnect cable this error means there is a communication loss to the technique processor board.
to rate and post a comment
Matt Du Vall
Re: ERROR 253 OEC 9600
March 25, 2012 11:08
This out of the manual
*Error 253 occurs when the system does not boot fully and displays the "system error
#253" message. This message is associated with a loss of serial communication.
Do you want a copy of the manual? reply to:
cbet_us@yahoo.com
to rate and post a comment
Matt Du Vall
Re: ERROR 253 OEC 9600
March 25, 2012 11:08
for the service manual,
google search: oec 9600 service manual it will be in a PDF
to rate and post a comment
Kenneth Calhoun
Re: ERROR 253 OEC 9600
April 12, 2012 10:37
the problem was a bad connection on the serial cable between the panel control board and Aux board.Their is a FSB# 94-020 on this problem.
to rate and post a comment
Matt Du Vall
Re: ERROR 253 OEC 9600
April 13, 2012 11:47
Cool.
to rate and post a comment