My VCT system reports these errors: 260100090 (NGPDU KSS open) and 260100093 (PDU HVDC Buss Fault Status). Jedi, tube and NGPDU Board are OK. I replaced KSS contactor, but I can't solve the problem. This NGPDU don't have installed the NGPDU RC filter cable kit (FMI 25442). Some suggestion to resolve the problem?
Hello Mr Saif. Yes, I resolved the problem. I found 2 of 3 resistances in the PDU damage. These resistances are at the transformer side. If you remove the PDU top cover you can see them to your left hand
Brand new OEM Philips & Siemens Parts at unbeatable prices shipping domestically and Includes warranty!
Email P/N for pricing: sales@Relaxationinc.com
(P) 813-925-1314, www.relaxationinc.com
VENGE
Error: 260100090
April 26, 2016 10:07
My VCT system reports these errors: 260100090 (NGPDU KSS open) and 260100093 (PDU HVDC Buss Fault Status). Jedi, tube and NGPDU Board are OK. I replaced KSS contactor, but I can't solve the problem. This NGPDU don't have installed the NGPDU RC filter cable kit (FMI 25442). Some suggestion to resolve the problem?
to rate and post a comment
SAIF ANSARI
re: Error: 260100090
August 10, 2020 05:49
sir, same i am facing this issue if it is resolved plz suggest me what was action taken by u
thank u
to rate and post a comment
VENGE
re: re: Error: 260100090
August 13, 2020 05:09
Hello Mr Saif. Yes, I resolved the problem. I found 2 of 3 resistances in the PDU damage. These resistances are at the transformer side. If you remove the PDU top cover you can see them to your left hand