Ampronix, a Top Master Distributor for Sony Medical, provides Sales, Service & Exchanges for Sony Surgical Displays, Printers, & More. Rely on Us for Expert Support Tailored to Your Needs. Email info@ampronix.com or Call 949-273-8000 for Premier Pricing.
EDGAR SANTIAGO RIVERA ZAMUDIO
ERROR 271 FCR PRIMA
September 13, 2016 08:17
Hello colleagues
I do not have the manual for this FCR PRIMA fijufilm
make the mistake
this error(271) marks when the image was already processed in the workstation
someone can help me that is familiar to this fault or the best alquien if someone has the service manual I hope can share...
Thankss...
Best Regards...
to rate and post a comment
ROBERTO DIAZ BEDOLLA
re: ERROR 271 FCR PRIMA
September 14, 2016 01:31
Correct sensitivity... you need to measure 10mR. So you need special equipment
to rate and post a comment
Gabriel Cool
re: ERROR 271 FCR PRIMA
September 14, 2016 01:36
Can you send me img of error and software version?
sarkisovkaren87@gmail.com
to rate and post a comment
EDGAR SANTIAGO RIVERA ZAMUDIO
re:re: ERROR 271 FCR PRIMA
September 15, 2016 09:40
Hi.
Thanks for your help
The version is V2.03.0000...
Regards
to rate and post a comment
Santiago Puerta Sánchez
re: re: ERROR 271 FCR PRIMA
May 21, 2019 09:26
Hello Gabriel, you can help me too? I have the same problem and I need the service manual.
Very thanks
to rate and post a comment
ROBERTO DIAZ BEDOLLA
re: ERROR 271 FCR PRIMA
September 15, 2016 02:17
Where are u located. Ask for fuji service. You need a special calibration. First you need a special xpose 14x17 ip to 10mR. So in order expose10mR you need to measure it first. If you are in Mexico We can help you
to rate and post a comment
ROBERTO DIAZ BEDOLLA
re: ERROR 271 FCR PRIMA
September 15, 2016 02:17
Email me