I disconnect and connect the x-ray tube and it worked again, but does not work with low values of MA and shows error code 32, could you help me with this error?
that's a mA sense error. Double-check you mA calibration with the scope. correct any undershoot or overshoot. also check the amplitude and be sure you have the correct mA amplitude.
36 CONTROL ERROR (Tube Rotor Fault)
‘System Error’
Symptom: “Rotor Okay” signal line low.
Action: Ohm fuses F2, F3, and F4 of the tube control board.
Ensure the x-ray tube is not overheated (x-ray tubes take approximately one hour to cool, they may be used
as soon as they have cooled sufficiently).
If you send me an email address I can send you some more info
Quite often I have seen one or more of these rotor fuses go bad. I usually change all three to be safe for the long haul. Definitely do an Ma preheat cal as outlined by Naelv. If you need a tube, I have an Elite that I use for parts that has a good tube on it.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Error 36 Lorad Elite
April 10, 2015 09:10
Hello friends, i have a trouble with a lorad Elite, the equipment showed an error: control error code 36, can anyone help me with this?
to rate and post a comment
Naelv
re: Error 36 Lorad Elite
April 11, 2015 09:03
it looks like you have a tube rotor problem. it could be the power to the rotor of the xray tube or the rotor inside the xray tube itself.
to rate and post a comment
Felipe Canche
re: re: Error 36 Lorad Elite
April 13, 2015 04:15
I disconnect and connect the x-ray tube and it worked again, but does not work with low values of MA and shows error code 32, could you help me with this error?
to rate and post a comment
Naelv
re: re: re: Error 36 Lorad Elite
April 14, 2015 10:03
that's a mA sense error. Double-check you mA calibration with the scope. correct any undershoot or overshoot. also check the amplitude and be sure you have the correct mA amplitude.
to rate and post a comment
Felipe Canche
re: re: re: re: Error 36 Lorad Elite
April 15, 2015 08:23
ok, thanks!
to rate and post a comment
Lenny Place
re: Error 36 Lorad Elite
April 13, 2015 04:16
36 CONTROL ERROR (Tube Rotor Fault)
‘System Error’
Symptom: “Rotor Okay” signal line low.
Action: Ohm fuses F2, F3, and F4 of the tube control board.
Ensure the x-ray tube is not overheated (x-ray tubes take approximately one hour to cool, they may be used
as soon as they have cooled sufficiently).
If you send me an email address I can send you some more info
to rate and post a comment
Larry Underriner
re: re: Error 36 Lorad Elite
April 14, 2015 04:36
Quite often I have seen one or more of these rotor fuses go bad. I usually change all three to be safe for the long haul. Definitely do an Ma preheat cal as outlined by Naelv. If you need a tube, I have an Elite that I use for parts that has a good tube on it.
to rate and post a comment
Felipe Canche
re: re: re: Error 36 Lorad Elite
April 15, 2015 08:23
thanks!