Error 36 Lorad Elite

Posted by an Anonymous User

Error 36 Lorad Elite

April 10, 2015 09:10

Hello friends, i have a trouble with a lorad Elite, the equipment showed an error: control error code 36, can anyone help me with this?

Log inor Register

to rate and post a comment

(28)
(6)

Naelv

re: Error 36 Lorad Elite

April 11, 2015 09:03

it looks like you have a tube rotor problem. it could be the power to the rotor of the xray tube or the rotor inside the xray tube itself.

Log inor Register

to rate and post a comment

(7)
(1)

Felipe Canche

re: re: Error 36 Lorad Elite

April 13, 2015 04:15

I disconnect and connect the x-ray tube and it worked again, but does not work with low values of MA and shows error code 32, could you help me with this error?

Log inor Register

to rate and post a comment

(28)
(6)

Naelv

re: re: re: Error 36 Lorad Elite

April 14, 2015 10:03

that's a mA sense error. Double-check you mA calibration with the scope. correct any undershoot or overshoot. also check the amplitude and be sure you have the correct mA amplitude.

Log inor Register

to rate and post a comment

(7)
(1)

Felipe Canche

re: re: re: re: Error 36 Lorad Elite

April 15, 2015 08:23

ok, thanks!

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Error 36 Lorad Elite

April 13, 2015 04:16

36 CONTROL ERROR (Tube Rotor Fault)
‘System Error’
Symptom: “Rotor Okay” signal line low.
Action: Ohm fuses F2, F3, and F4 of the tube control board.
Ensure the x-ray tube is not overheated (x-ray tubes take approximately one hour to cool, they may be used
as soon as they have cooled sufficiently).

If you send me an email address I can send you some more info

Log inor Register

to rate and post a comment

(28)
(1)

Larry Underriner

re: re: Error 36 Lorad Elite

April 14, 2015 04:36

Quite often I have seen one or more of these rotor fuses go bad. I usually change all three to be safe for the long haul. Definitely do an Ma preheat cal as outlined by Naelv. If you need a tube, I have an Elite that I use for parts that has a good tube on it.

Log inor Register

to rate and post a comment

(7)
(1)

Felipe Canche

re: re: re: Error 36 Lorad Elite

April 15, 2015 08:23

thanks!

Log inor Register

to rate and post a comment