Error 411 on GE MPH50

November 01, 2020 11:05

Hi,
do you have information about the mph50 generator error 411, FPS_FAULT signal; status error?
The voltage is symmetrical and correct, but error 411 appears first and then 513.

1 person liked this
Log inor Register

to rate and post a comment

(102)
(39)
(1)

Konstantin Kuznetsov

re: Error 411 on GE MPH50

November 02, 2020 10:21

MPH Code: 411
Internal Code: FPS_ERROR
Cause: Inverter floating power supply
default (fps_fault signal)
Effect: Exposure inhibited
Action:Reset error.
If it reappears, find the default in running diagnostics
on/off and DC bus
MENU2 TEST F1
MENU2 TEST F3
Then follow the faulty tree ON/OFF

MPH Code: 513
Internal Code: HE_INVERTER
Cause: Inverter safety is set out of exposure (safety, kv_inf_kv0,
exp_time_max, restarting_safety, kv_drop,
kv_drop_an, kv_drop_cat, kv_max, mA_max, regul_out,
i_inv_max, kv_sup_75kvn signals)
Effect: Exposure inhibited
Action: Reset error.
If it reappears, find the default in running diagnostic kV
MENU2 TEST F6
MENU2 TEST F7

The service manual is available in the Documentation Library / Customer Documentation Portal on the GE Healthcare official website. Document number 2165118-1EN. There is generally a lot of things available)

Log inor Register

to rate and post a comment

(8)
(2)

Alberto Pepe

re: re: Error 411 on GE MPH50

November 06, 2020 11:01

Hi Konstantin,
I have the service manual and the schematics, but I don't have the service software to check.
From the measurements made, there are no discrepancies on the tensions, they are stable and symmetrical.
Besides that, I have problems on the ARC2000 starter, of which I am missing the schematics, can you help me?

Log inor Register

to rate and post a comment

(1)

eusebio rocha

re: Error 411 on GE MPH50

November 02, 2020 10:23

i have no't the manual for this equipment, I sorry.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)