Error 6C4A2 on AGFA Dr3000

(4)

JAOZAFY Aly Johnny

Error 6C4A2 on AGFA Dr3000

January 27, 2012 10:21

Erreur 6C4A2 apparaît au démarrage de l'imprimante pendant la séquence d'initialisation.

Carte AIC1 Rev -H - N/S 03230H :
Led1 Led2 Led3 VCC -VS +Vs
OFF OFF ON ON ON ON
Vth PWM
ON ON


Carte PMC1 Rev H - N/S 304K0447
L1 L2 L3 L4 L5 L6
ON OFF Flash Flash Flash ON
L7 Heater +5V +8V +12V -12V
ON OFF ON ON ON ON
+24V
ON


GEMINI Board
OFF OFF OFF
ON Flash OFF

OFF OFF OFF
Flash OFF OFF


COS Board
vert jaune Rouge
OFF OFF OFF
OFF OFF OFF
OFF ON ON

Led2 Led5 Led6 Led7
OFF OFF OFF OFF


ALIMENTATION :
Les 6 Leds vertes = ON

COMMENT FAIRE?

Log inor Register

to rate and post a comment

Jose Victor Perez Losada

Re: Error 6C4A2 on AGFA Dr3000

January 27, 2012 05:35

Hello, please contact me in vplosada@gmail.com
Regards

Log inor Register

to rate and post a comment

(192)
(45)
(1)

Dante Nuevo

Re: Error 6C4A2 on AGFA Dr3000

January 30, 2012 10:23

Error Code --> 6c4a2
E_T14_VMEAS_THREF,
Meaning : Error in determining min/max of power-supply.
Difference between min/max-read-out of
power-supply is too small.
Reason : a. ADC-cycle fails.
b. Connector allowing PWM-voltage control
is broken or not inserted.
Cure : a. Check resource-status -> AIC-board.
b. Check cable(s) to power-supply.


I hope it helps you.

Dante Nuevo