Hello. There was an error 3 appearing at the back of Acuson Aspen. What am I supposed to do with that error? I did component level repair on main dc power supply. There were no defective components. What are the other solutions? Thanks and God bless!

Log inor Register

to rate and post a comment

Rick Reed

re: Error code 3 on Acuson Aspen

September 26, 2012 09:07

System shuts down with an Error Code “3”
displayed (i.e., Main Power Supply CODC overcurrent).
One or more of the DC power supplies is shorted to ground, or to another DC voltage. Check for power supply shorts by measuring the impedance of appropriate power supply outputs with the system off and AC power
disconnected.

Log inor Register

to rate and post a comment

(33)
(3)

Joao Carlso

re: Error code 3 on Acuson Aspen

September 26, 2012 10:09

Dear Mendoza:
Check for shorted capacitors on the boards, not the source.
I have found capacitors short in several different boards, are usually those who are at the entrance the tensions to the plates.
Good luck

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

re: Error code 3 on Acuson Aspen

September 27, 2012 07:16

e-mail me I have the service manual crisb1955@aol.com

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)