Hello colleagues,
please share your experience with GE CT/e Dua! Have you changed an Х-ray tube after a 30-1010-01 - Zero detect error? CAM, CIF, DDP, DAS Power supply (adjusted and checked for ripples and noise), have already been reviewed.
Thanks in advance!
There is a switch named SW3 on CIF. Board. Change its its switch no 8 to on position . By doing this the zero das data error will no appear . And it will help to diagnose a faulty cam board
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Mario Icherenski
Error Code: 30-1010-01
August 12, 2022 10:53
Hello colleagues,
please share your experience with GE CT/e Dua! Have you changed an Х-ray tube after a 30-1010-01 - Zero detect error? CAM, CIF, DDP, DAS Power supply (adjusted and checked for ripples and noise), have already been reviewed.
Thanks in advance!
to rate and post a comment
Jogi
re: Error Code: 30-1010-01
August 15, 2022 10:18
There is a switch named SW3 on CIF. Board. Change its its switch no 8 to on position . By doing this the zero das data error will no appear . And it will help to diagnose a faulty cam board